9.2.2. PERFORMING A BASIC MANUAL CALIBRATION CHECK

Analyzer display
continues to cycle
through all of the
available gas
measurements
throughout this
procedure.
Wait until STABIL falls
below 0.2 PPM (for T300)
or 1.0 PPM (for T300M).
This may take several
minutes.
06864B DCN6314
SAMPLE
RANGE=50.0PPM
< TST
TST >
CAL
Toggle TST> button until ...
SAMPLE
STABIL= XXXX PPM
< TST TST >
CAL
Allow zero gas to enter the sample port
at the rear of the analyzer.
Record CO zero point readings
Allow span gas to enter the sample port
at the rear of the analyzer.
Record CO span point readings\
CO=XX.XX
MSG
SETUP
CO=XX.XX
SETUP
Calibration Procedures
Set the Display to show
the STABIL test function.
This function calculates
the stability of the CO
measurement.
Wait until STABIL falls
below or 0.2 PPM (for
T300) or 1.0 PPM (for
T300M).
This may take several
minutes.
The ZERO and/or SPAN
keys will appear at various
points of this process.
It is not necessary to press
them.