Teledyne API – Model T300/T300M CO Analyzer
DISPLAYED AS
= CHANNEL
CONC_OUT_1
=
A1
CONC_OUT_2
=
A2
CONC_OUT_3
=
A3
TEST OUTPUT
=
A4
CONC_OUT_3
is only active on
analyzers with the
optional CO
or O
2
sensor installed.
These buttons increase /
decrease the analog output
signal level (not the value on the
display)
by 100, 10 or 1 counts.
Continue adjustments until the
voltage measured at the output
of the analyzer and/or the input
of the recording device matches
the value in the upper right hand
corner of the display
(within the tolerances
listed in Table 8.7.)
06864B DCN6314
To adjust the signal levels of an analog output channel manually, select the ANALOG
I/O CONFIGURATION submenu (see Figure 5-3) then press:
DIAG
PREV NEXT
DIAG AIO
SET>
Continue pressing SET> until you reach the
DIAG AIO
<SET SET>
DIAG AIO
2
DIAG AIO
<SET SET>
DIAG AIO
U100 UP10 UP
DIAG AIO
U100 UP10 UP
DIAG AIO
<SET SET> CAL
From the
AIO CONFIGURATION SUBMENU
ANALOG I/O CONFIGURATION
ENTR
AOUTS CALIBRATED: NO
CAL
output to be configured
CONC_OUT_2: 5V, CONC2, NOCAL
EDIT
CONC_OUT_2: RANGE: 5V
SET>
EDIT
Continue pressing SET> until ...
CONC_OUT_2: CALIBRATED:NO
CAL
CONC_OUT_2: VOLT-Z: 0 mV
DOWN DN10 D100 ENTR
CONC_OUT_2: VOLT-S: 4500 mV
DOWN DN10 D100 ENTR EXIT
CONC_OUT_2: CALIBRATED: YES
Setup Menu
EXIT
EXIT
EXIT
EXIT
EXIT
EXIT
These menus
only appear if
AUTO-CAL is
turned OFF.
EXIT
129