Individual Calibration Of The Analog Outputs - Teledyne T300 Operation Manual

Carbon monoxide analyzer
Table of Contents

Advertisement

Teledyne API – Model T300/T300M CO Analyzer

5.9.3.5. INDIVIDUAL CALIBRATION OF THE ANALOG OUTPUTS

06864B DCN6314
To use the AUTO CAL feature to initiate an automatic calibration for an individual
analog output, select the ANALOG I/O CONFIGURATION submenu (see Figure 5-3)
then press:
AIO CONFIGURATION SUBMENU
DIAG
ANALOG I/O CONFIGURATION
PREV NEXT
DIAG AIO
AOUTS CALIBRATED: NO
SET>
CAL
Continue pressing SET> until you reach the
output to be configured
DIAG AIO
CONC_OUT_2: 5V, CONC2, NOCAL
<SET SET>
EDIT
DIAG AIO
CONC_OUT_2: RANGE: 5V
SET>
EDIT
Continue pressing SET> until ...
DIAG AIO
CONC_OUT_2: CALIBRATED:NO
<SET SET>
CAL
DIAG AIO
AUTO CALIBRATING CONC_OUT_2
DIAG AIO
CONC_OUT_2: CALIBRATED: YES
<SET SET> CAL
From the
ENTR
EXIT
EXIT
EXIT
EXIT
EXIT
EXIT
Setup Menu
127