Calibration And Check Procedures For Basic Configuration; Figure 4-1. Multi-Point Calibration Page - Teledyne T200 User Manual

No/no2/nox analyzer with numaview software
Hide thumbs Also See for T200:
Table of Contents

Advertisement

Note
To calibrate or to perform a calibration check for basic configuration instruments, see
Section 4.2.1.
To calibrate or to perform a calibration check for instruments with valve options, see
Section 4.2.2.
To perform a calibration check for instruments with the internal span gas generator, see
Section 4.3.
To calibrate Conversion Efficiency, see Section 4.5.
C
ALIBRATION AND
Although this section uses the Calibration menu for both check and actual calibration, a
check does not require the Calibration menu. Instead, while in Home page, simply flow the
zero air or the NO span gas through the Sample port, and check the reading after the
Stability falls below 1.0 PPB (either in the gas graph or in the Dashboard).
Otherwise, follow the steps presented in Sections 4.2.1.1 and 4.2.1.2.
96
Teledyne API T200 NO/NO
Tips for Setting the Expected Span Gas Concentration:
• The expected concentrations for both NO and NOx are usually
set to the same value.
• When setting expected concentration values, consider
impurities in your span gas source (e.g. NO often contains 1-
3% NO
and vice versa).
2
• The NO and NOx span gas concentrations should be 80% to
90% of range of concentration values likely to be encountered
in your application.
C
P
HECK
ROCEDURES FOR

Figure 4-1. Multi-Point Calibration Page

/NO
Analyzer with NumaView™ Software
2
x
B
C
ASIC
ONFIGURATION
083730200A DCN7962