TROUBLESHOOTING and REPAIR
4. Finger tighten the thumbscrew.

11.5.1.2. Replacing the UV Lamp

DO NOT grasp the UV lamp by its cap when changing its position (see Figure 10-1).
72
DO NOT over-tighten the thumbscrew.

Figure 11-1. Shutter Assembly - Exploded View

1.
Turn off the analyzer.
2.
Disconnect the UV lamp from its power supply.
You can find the power supply connector by following the two UV Lamp
power supply wires from the lamp to the power supply.
3.
Loosen, but do not remove the two UV lamp bracket screws, and the
large brass thumbscrew located on the shutter housing (see Figure 10-
1) so that the lamp can be moved.
Always grasp the main body of the lamp.
4.
Remove the UV Lamp by pulling it straight up.
5.
Insert the new UV lamp into the bracket.
6.
Tighten the two UV lamp bracket screws, but leave the brass thumb
screw un-tightened.
7.
Connect the new UV lamp to the power supply.
8.
Turn the instrument on and perform the UV adjustment procedure as
defined in Section 11.5.1.1of this addendum
9.
Finger tighten the thumbscrew.
DO NOT over-tighten the thumbscrew.
Teledyne API - T102/501 TRS, Addendum to T101 Operation Manual