Teledyne API – Model T700 Dynamic Dilution Calibrator

4.11.6. CONFIGURING LEVEL STATUS BLOCKS

Make sure that the T700 is
in standby mode.
Toggle these buttons
until the number of the
LEVEL to be edited is
reached.
Toggle this button
to turn the
STATUS BLOCK
ON/OFF
Moves the
cursor one
character left or
right.
06873E DCN7106
There are two STATUS BLOCKS associated with LEADS LEVELS.
OUTPUT connections located on the back panel of the T700 (see Figure 3-4 and
Section 3.3.1.5).
physical output but is used to communicate status over the serial data port.
To configure the either of the STATUS BLOCKS, press:
STANDBY
<TST
TST>
SETUP X.X
GAS SEQ
LVL
SETUP X.X
[LEVEL ID] )[Gas/Conc.],[Status Block Set'g]
PREV NEXT
SETUP X.X
<SET SET>
Continue pressing SET> until Desired
SETUP X.X
STATUS BLOCK 2:DISABLED
<SET SET>
SETUP X.X
STATUS BLOCK 2:OFF
OFF
SETUP X.X
STATUS BLOCK 2:[0]00000000000
<CH
CH>
Toggle this button to turn the selected bit ON/OFF (0 or 1).
Each bit shown on the display represents one of the control
output pins located on the back of the M700E (see Figure 3-2),
The left most bit is Bit 1, the next bit to the right, bit 2,
progressing rightward to bit 12 (see Figure 3-10 for connector
Overview of Operating Modes and Basic Operation
BLOCK 1: This block corresponds to the physical CONTROL
BLOCK 2: The second status block does not correspond to any
A- CAL=0.000 LPM
GEN STBY SEQ
PRIMARY SETUP MENU
CFG
CLK PASS
MORE
INS
PRNT
DEL EDIT
LEVEL NUMBER:12
EDIT
Status Block is reached
EDIT
ENTER EXIT
[0]
ENTER EXIT
pin assignments)
SETUP
EXIT
EXIT
EXIT
EXIT
EXIT discards the
new setting
ENTR accepts the
new setting
EXIT discards the
new setting
ENTR accepts the
new setting
169