Selecting A Test Channel Function To Output; Table 4-14: Test Channels Functions Available On The T700"S Analog Output - Teledyne T700 Operation Manual

Dynamic dilution calibrator
Hide thumbs Also See for T700:
Table of Contents

Advertisement

Overview of Operating Modes and Basic Operation

4.10.1.2. Selecting a Test Channel Function to Output

Table 4-14: Test Channels Functions available on the T700's Analog Output
TEST CHANNEL
NONE
O3 PHOTO MEAS
O3 PHOTO REF
REGULATOR PRESSURE
SAMPLE PRESSURE
SAMPLE FLOW
SAMPLE TEMP
PHOTO LAMP TEMP
O3 LAMP TEMP
CHASSIS TEMP
O3 PHOTO CONC
150
The Test Functions available to be reported are listed on Table 4-14:
The raw output of the photometer during its
The raw output of the photometer during its
The gas pressure of the pressure regulator
on the O3 generator supply line
The pressure of gas in the photometer
The gas flow rate through the photometer
The temperature of gas in the photometer
The temperature of the photometer UV lamp
The temperature of the O
The temperature inside the T700‟s chassis
(same as BOX TEMP)
The current concentration of O
measured by the photometer.
Once a function is selected, the instrument not only begins to output a signal on the
analog output, but also adds TEST to the list of Test Functions viewable via the Front
Panel Display.
Teledyne API – Model T700 Dynamic Dilution Calibrator
DESCRIPTION
TEST CHANNEL IS TURNED OFF
measure cycle
reference cycle
absorption tube
absorption tube
generator‟s UV
3
lamp
being
3
ZERO
FULL SCALE
0 mV
5000 mV
0 mV
5000 mV
PSIG
PSIG
0" Hg-In-
40" Hg-In-A
A
3
3
0 cm
/min
1000 cm
/min
0 C
70 C
0 C
70 C
0 mV
5000 mV
0 C
70 C
0 PPM
1 ppm
06873E DCN7106