Teledyne API – Model T700 Dynamic Dilution Calibrator
4.3.2.2. The GPT Step
Toggle these buttons
to set the NO target
concentration.
MUST be at least 10%
Higher than the Target
O
Concentration
3
Toggle these buttons
to set the NO target
concentration.
Should be equal to the
expected NO
2
concentration
Toggle these buttons to
set the target TOTAL
FLOW.
(Default = 2.000 LPM)
Toggle these
buttons to set
DURATION of this
step
06873E DCN7106
This step operates and is programmed similarly to the GENERATE >GPT (see Section
4.2.3 for information on choosing the correct input values for this step).
At the end of the programming sequence, the T700 firmware will automatically insert a
DURATION step that needs to be defined.
To insert a GPT step into a sequence, press:
Starting at the STEPS Submenu
INSERT STEP Submenu
SETUP X.X
INSERT STEP: GENERATE
PREV NEXT
Use the PREV and NEXT keys to scroll though the
list of available instructions
SETUP X.X
INSERT STEP: GPT
PREV NEXT
GPT
GPT:0.0 PPB NO
0
0
0
GPT
GPT:0.0 PPB O3
0
0
0
GPT
TOTAL FLOW = 2.000 LPM
0
2.
0
SETUP X.X
DURATION: 10.0 MIN
0
1
0
SETUP X.X
3) DURATION : 10.0 MIN
PREV NEXT
Overview of Operating Modes and Basic Operation
EXIT
ENTR
EXIT
ENTR
.0
PPB
ENTR EXIT
.0
PPB
ENTR EXIT
0
0
ENTR EXIT
.0
ENTR
EXIT
INS
DEL EDIT
EXIT
EXIT discards the new gas type &
target concentration
ENTR accepts the new gas type &
target concentration
EXIT discards the new gas type &
target concentration
ENTR accepts the new gas type &
target concentration
EXIT discards the new
flow rate
ENTR accepts the
new gas flow rate
EXIT discards the new
setting
ENTR accepts the
new setting
129