Naming A Sequence - Teledyne T700 Operation Manual

Dynamic dilution calibrator
Hide thumbs Also See for T700:
Table of Contents

Advertisement

Overview of Operating Modes and Basic Operation

4.3.1.2. Naming a Sequence

Note
This display only appears if there are no sequences currently
programmed into the T700.
OTHERWISE ...
SETUP X.X
1) SEQ [NAME], [X] STEPS
PREV NEXT
Deletes the sequence shown
in the message field
Scrolls back and forth between
existing sequences
118
The first step of creating a calibration sequence is to assign it a name. The name can be
up to 10 characters and can be comprised of any alpha character (A to Z), and numeral
(0 to 9) or the underscore character ("_").
All sequences should be assigned a unique name.
To assign a name to a sequence, press:
Make sure that the T700 is
in standby mode.
INS
PRNT
EXIT
DEL EDIT
Edits the sequence shown
in the message field
Moves the
cursor one
character left or
right.
Teledyne API – Model T700 Dynamic Dilution Calibrator
STANDBY
A-CAL=0.000 LPM
<TST
TST>
GEN STBY SEQ
SETUP X.X
PRIMARY SETUP MENU
GAS
CFG
CLK PASS
SEQ
SETUP X.X
SEQUENCE CONFIGURATION
PRINT
EDIT
SETUP X.X
END OF SEQUENCES
INS
SETUP X.X
NAME:0
SET>
EDIT
SETUP X.X
NAME:[0]
<CH
CH>
INS
DEL
[0]
Inserts a new a
character at the
Deletes a
cursor location.
character at the
cursor location.
SETUP
MORE
EXIT
EXIT
PRNT
EXIT
EXIT
EXIT discards the
ENTER EXIT
new NAME
ENTR accepts the
new NAME
Toggle this button to
cycle through the range of
numerals and available
characters:
("A – Z"; "0 – 9" & " _ ")
06873E DCN7106

Advertisement

Table of Contents
loading

Table of Contents