Performing An Automatic Calibration Of The Optional O Generator - Teledyne T700 Operation Manual

Dynamic dilution calibrator
Hide thumbs Also See for T700:
Table of Contents

Advertisement

Calibration and Verification
7.4.2.5. Performing an Automatic Calibration of the Optional O
Note
Make sure that the T700 is
in standby mode.
Toggle these buttons to
enter the correct
PASSWORD
214
This procedure requires that the T700 calibrator have an optional photometer
installed.
To run the automatic O
STANDBY
<TST
TST>
SETUP X.X
GAS
SEQ
SETUP X.X
COMM FLOW
SETUP X.X
8
1
DIAG
PREV
NEXT
Continue pressing NEXT until ...
DIAG
PREV NEXT
DIAG
CAL
PNTS
DIAG
DIAG
Teledyne API – Model T700 Dynamic Dilution Calibrator
generator calibration program, press:
3
A-CAL=0.000 LPM
GEN STBY SEQ
PRIMARY SETUP MENU
CFG
CLK PASS
MORE
SECONDARY SETUP MENU
VAr
DIAG
ENTER PASSWORD
8
SIGNAL I/O
O3 GEN CALIBRATION
ENTR
O3 GEN CALIBRATION
O3 GEN CAL 0% COMPLETE
Test runs automatically
O3 GEN CAL 100% COMPLETE
Generator
3
SETUP
EXIT
EXIT
ENTR
EXIT
ENTR
EXIT
EXIT
EXIT
EXIT
EXIT aborts
the calibration
06873E DCN7106