Deleting Or Editing An Individual Step In A Sequence - Teledyne T700 Operation Manual

Dynamic dilution calibrator
Hide thumbs Also See for T700:
Table of Contents

Advertisement

Overview of Operating Modes and Basic Operation

4.3.2.10. Deleting or Editing an Individual Step in a Sequence

Make sure that the
T700 is in standby
mode.
Scrolls back and forth
between existing
sequences
If the step deleted was GENERATE,
MANUAL or PURGE, delete the
following duration step
136
To delete or edit an individual step in an existing Sequence, press:
STANDBY
<TST
TST>
GEN STBY SEQ
SETUP X.X
PRIMARY SETUP MENU
GAS
SEQ
CFG
SETUP X.X
SEQUENCE CONFIGURATION
EDIT
PRINT
SETUP X.X
PREV NEXT
SETUP X.X
NAME:0
SET>
EDIT
Continue pressing SET> until ...
STEPS Submenu
SETUP X.X
STEPS: 12
<SET SET>
EDIT
SETUP X.X
1) STANDBY
PREV NEXT
Continue pressing NEXT or PREV until until
step to be deleted appears
SETUP X.X
7) PURGE
PREV NEXT
SETUP X.X
DELETE STEP?
YES
NO
SETUP X.X
7) DURATION : 10.0 MIN
PREV NEXT
Teledyne API – Model T700 Dynamic Dilution Calibrator
ACT CAL=0.000 LPM
SETUP
CLK PASS
MORE
[NAME]
[X]
1) SEQ
,
STEPS
INS
DEL EDIT
PRNT
INS
DEL
INS
DEL EDIT
INS
DEL
EDIT
EXIT
EXIT
EXIT
EXIT
EXIT
EXIT
EXIT
Follow the
EXIT
instructions for
the step being
edited
06873E DCN7106