Calibration and Verification
7.4.2. O
GENERATOR CALIBRATION PROCEDURE
3
7.4.2.1. Viewing O
3
Make sure that the T700 is
in standby mode.
Toggle these buttons to
enter the correct
PASSWORD
Toggle these buttons to move
between calibration points
210
The T700 calibrator‟s software includes a routine for automatically calibrating the O
generator. A table of drive voltages stored in the T700‟s memory is the basis for this
calibration. For each point included in the table used by the T700 to calibrate the
optional O
generator the user can set a drive voltage and a dwell time for that point.
3
Each point can also be individually turned OFF or ON.
Generator Calibration Points
To view these calibration points, press:
STANDBY
<TST
SETUP X.X
GAS
SETUP X.X
COMM FLOW
SETUP X.X
DIAG
PREV
DIAG
PREV NEXT
DIAG
CAL
DIAG O3GEN
PREV NEXT
Teledyne API – Model T700 Dynamic Dilution Calibrator
A-CAL=0.000 LPM
TST>
GEN STBY SEQ
PRIMARY SETUP MENU
SEQ
CFG
CLK PASS
SECONDARY SETUP MENU
VARS
DIAG
ENTER PASSWORD
8
1
8
SIGNAL I/O
NEXT
Continue pressing NEXT until ...
O3 GEN CALIBRATION
O3 GEN CALIBRATION
PNTS
1) 500 MV, 5.0 MIN, ON
INS
SETUP
EXIT
MORE
EXIT
EXIT
ENTR
ENTR
EXIT
EXIT
ENTR
EXIT
DEL EDIT
PRNT
EXIT
3
06873E DCN7106

Advertisement

Table of Contents
loading

Table of Contents