3.4. STARTUP, FUNCTIONAL CHECKS, AND INITIAL
CALIBRATION

3.4.1. START UP

74
If you are unfamiliar with the T700 principles of operation, we recommend that you read
Section 0.
For information on navigating the calibrator‟s software menus, see the menu trees
described in Appendix A.
After the electrical and pneumatic connections are made, an initial functional check is in
order. Turn on the instrument.
installed) should start immediately. The front panel display will show a splash screen
and other information during the initialization process while the CPU loads the operating
system, the firmware and the configuration data.
The calibrator should automatically switch to STANDBY mode after completing the
brief boot-up sequence. However, the T700 dynamic dilution calibrator requires a
minimum of 30 minutes for all of its internal components to reach a stable operating
temperature. During the warm-up period, the front panel display may show messages in
the Parameters field.
Teledyne API – Model T700 Dynamic Dilution Calibrator
The exhaust fan (and pump if photometer option
– Special)
y
06873E DCN7106