4.10.1.3. Test Channel Voltage Range Configuration

RANGE SPAN
These buttons
set the signal
level and type
of the selected
channel
152
In its standard configuration, the analog outputs is set to output a 0 – 5 VDC signals.
Several other output ranges are available (see Table 4-15). Each range is usable from -
5% to + 5% of the rated span.

Table 4-15: Analog Output Voltage Range Min/Max

MINIMUM OUTPUT
0-100 mVDC
0-1 VDC
0-5 VDC
0-10 VDC
The default offset for all ranges is 0 VDC.
To change the output range, press:
From the
AIO CONFIGURATION SUBMENU
(See Section 6.9.1.1)
DIAG
ANALOG I/O CONFIGURATION
PREV NEXT
DIAG AIO
AOUTS CALIBRATED: NO
CAL
SET>
Continue pressing SET> until you reach the
output to be configured
DIAG AIO
TEST_OUTPUT: 5V, OVR, NOCAL
<SET SET>
EDIT
DIAG AIO
TEST_OUTPUT: RANGE: 5V
0.1V
1V
5V
10V
Teledyne API – Model T700 Dynamic Dilution Calibrator
MAXIMUM OUTPUT
-5 mVDC
-0.05 VDC
-0.25 VDC
-0.5 VDC
ENTR
ENTR EXIT
105 mVDC
1.05 VDC
5.25 VDC
10.5 VDC
EXIT
EXIT
Pressing ENTR records
the new setting and
EXIT
returns to the previous
menu.
Pressing EXIT ignores the
new setting and returns to
the previous menu.
06873E DCN7106