Page 3
OLLUTION NSTRUMENTATION Teledyne Advanced Pollution Instrumentation (TAPI), a business unit of Teledyne Instruments, Inc., is a worldwide market leader in the design and manufacture of precision analytical instrumentation used for air quality monitoring, continuous emissions monitoring, and specialty process monitoring applications. Founded...
Page 4
Teledyne API – Model T700 Dynamic Dilution Calibrator This page intentionally left blank. 06873B DCN6388...
Page 5
NEVER use any gas analyzer to sample combustible gas(es)! For Technical Assistance regarding the use and maintenance of this Note instrument or any other Teledyne API product, please contact Teledyne API’s Customer Service Department: Telephone: 800-324-5190 Email: api-customerservice@teledyne.com or by accessing various service options on our website: http://www.teledyne-api.com/.
Page 6
Teledyne API – Model T700 Dynamic Dilution Calibrator CONSIGNES DE SÉCURITÉ Des consignes de sécurité importantes sont fournies tout au long du présent manuel dans le but d’éviter des blessures corporelles ou d’endommager les instruments. Veuillez lire attentivement ces consignes. Chaque consigne de sécurité...
Page 7
PRODUCT RETURN All units or components returned to Teledyne API should be properly packed for handling and returned freight prepaid to the nearest designated Service Center. After the repair, the equipment will be returned, freight prepaid.
Page 8
Teledyne API – Model T700 Dynamic Dilution Calibrator This page intentionally left blank. 06873B DCN6388...
Page 9
Presented here is information regarding the documents that are included with this manual (Structure) and how the content is organized (Organization). STRUCTURE This T700 manual, PN 06873, is comprised of multiple documents, assembled in PDF format, as listed below. Part No.
Page 10
Teledyne API – Model T700 Dynamic Dilution Calibrator and a Glossary, and also has a section that provides important information about electro- static discharge and avoiding its consequences. The appendices at the end of this manual provide support information such as, version- specific software documentation, lists of spare parts and schematics.
Page 11
REVISION HISTORY This section provides information regarding the initial release and subsequent changes to this manual. T700 Manual, PN 06873 Rev B, DCN 6388 Date Change Summary 2012 May 08 6388 Administrative changes: restructure/reformat Technical changes: various 2010 October 06...
Page 12
Teledyne API – Model T700 Dynamic Dilution Calibrator This page intentionally left blank. 06873B DCN6388...
3.3.2.1. About Diluent Gas (Zero Air) ......................54 3.3.2.2. About Calibration Gas ........................54 3.3.2.3. Connecting Diluent Gas to the Calibrator..................58 3.3.2.4. Connecting Calibration Source Gas to the T700 Calibrator ............58 3.3.2.5. Connecting Gas Outputs from the Calibrator .................59 3.3.2.6. Other Pneumatic Connections .......................63 3.3.3.
Page 14
4.2.4.1. T700 Calibrator GPTPS Operation....................113 4.2.4.2. Initiating a GPT Pre-Set....................... 115 4.2.5. GENERATE PURGE: Activating the T700’s Purge Feature ............116 4.2.6. GENERATE ACT>: VIEWING CONCENTRATIONS Generated from Multi-Gas Cylinders... 118 4.2.6.1. Using the T700 Calibrator as an O Photometer.................
Page 15
7.1. Viewing the Performance Statistics for the T700’s MFC’s............... 197 7.2. Calibrating the Output of the T700’s MFC’s....................199 7.2.1. Setup for Verification and Calibration of the T700’s MFC’s..............200 7.2.2. Verifying and Calibrating the T700’s MFC’s ..................201 7.3. Verifying and Calibrating the T700’s Optional O Photometer..............
Page 16
8.2.1.2. Setup Auto Leak Check....................... 227 8.2.1.3. Performing the Auto Leak Check Procedure................230 8.2.1.4. Returning the T700 to Service after Performing an Auto Leak Check ........230 8.2.2. Cleaning or Replacing the Absorption Tube..................231 8.2.3. UV Source Lamp Adjustment ......................232 8.2.4.
Page 18
11.4.2.2. Working at an Anti-ESD Work Bench..................310 11.4.2.3. Transferring Components from Rack to Bench and Back............311 11.4.2.4. Opening Shipments from Teledyne API’s Customer Service............ 311 11.4.2.5. Packing Components for Return to Teledyne API’s Customer Service ........312 06873B DCN6388...
Page 19
Figure 3-19: Set up for T700 – Connecting the T700 to a Calibration Manifold............61 Figure 3-20: Set up for T700 – Connecting the T700 to a Dual Span Gas / Zero Air Manifold ......62 Figure 3-21: T700 Pneumatic Diagram – Base Unit....................63 Figure 3-22: Internal Pneumatics for T700 Calibrator with Optional O Generator and GPT Chamber.
Page 20
Teledyne API – Model T700 Dynamic Dilution Calibrator Figure 8-2: Gas Port Setup for Auto-Leak Check Procedure................228 Figure 8-3: Gas Flow for Auto-Leak Check Procedure of Base Model T700’s ............ 229 Figure 8-4: Gas Flow for Auto-Leak Check Procedure of T700’s with Optional Photometer....... 229 Figure 8-5: Photometer Assembly –...
Page 21
Table 4-12: Variable Names (VARS) ........................148 Table 4-13: DIAG – Analog I/O Functions......................151 Table 4-14: Test Channels Functions available on the T700’s Analog Output ............ 154 Table 4-15: Analog Output Voltage Range Min/Max.................... 156 Table 4-16: Voltage Tolerances for the TEST CHANNEL Calibration ..............163 Table 5-1: COMM Port Communication Modes ....................
Page 22
Table 9-11: Analog Output Test Function – Nominal Values Voltage Outputs ............ 259 Table 9-12: Status Outputs Check ........................260 Table 9-13: T700 Control Input Pin Assignments and Corresponding Signal I/O Functions ....... 261 Table 9-14: Control Outputs Pin Assignments and Corresponding Signal I/O Functions Check ......262 Table 10-1: Relay PCA Status LEDs........................
INTRODUCTION This section provides an overview of the Model T700 calibrator, its features, and its options. 1.1. T700 CALIBRATOR OVERVIEW The Model T700 (typically referred to as T700) is a microprocessor-controlled calibrator for precision gas calibrators. Using a combination of highly accurate mass flow controllers and compressed sources of standard gases, calibration standards are provided for multipoint span and zero checks.
To order these options or to learn more about them, please contact the Sales Department of Teledyne Advanced Pollution Instruments at: TOLL-FREE:...
Page 27
Number CAUTION – GENERAL SAFETY HAZARD THE T700 CALIBRATOR WEIGHS CLOSE TO 18 KG (40 POUNDS) WITH ALL OPTIONS. TO AVOID PERSONAL INJURY WE RECOMMEND THAT TWO PERSONS LIFT AND CARRY IT BUT FIRST DISCONNECTING ALL CABLES AND TUBING FROM THE CALIBRATOR BEFORE MOVING IT.
Page 28
Introduction Teledyne API – Model T700 Dynamic Dilution Calibrator This page intentionally left blank. 06873B DCN6388...
CAL Gas Input Ports 4 (configurable) Diluent Gas Input Ports Response Time 60 Seconds (98%) Table 2-2: T700 Dilution Electrical and Physical Specifications Parameter Specification AC Power 85V-264V, 47Hz-63Hz At 115V ~ Start up: 110 W, Steady State: 140 W...
<1% / 24 hours Minimum Gas Flow Required 800 cc/min 2.2. APPROVALS AND CERTIFICATIONS The Teledyne API Model T700 calibrator was tested and certified for Safety and Electromagnet Compatibility (EMC). This section presents the compliance statements for those requirements and directives. 2.2.1. SAFETY...
Teledyne API – Model T700 Dynamic Dilution Calibrator Specifications and Approvals 2.2.2. EMC EN 61326-1 (IEC 61326-1), Class A Emissions/Industrial Immunity EN 55011 (CISPR 11), Group 1, Class A Emissions FCC 47 CFR Part 15B, Class A Emissions CE: 2004/108/EC, Electromagnetic Compatibility Directive 2.2.3.
Page 32
Specifications and Approvals Teledyne API – Model T700 Dynamic Dilution Calibrator This page intentionally left blank. 06873B DCN6388...
3.1. UNPACKING AND INITIAL SETUP CAUTION – RISK of Personal Injury THE T700 WEIGHS ABOUT 18 KG (40 POUNDS) WITHOUT OPTIONS INSTALLED. TO AVOID PERSONAL INJURY, WE RECOMMEND USING TWO PERSONS TO LIFT AND CARRY THE CALIBRATOR.
Getting Started Teledyne API – Model T700 Dynamic Dilution Calibrator WARNING! NEVER DISCONNECT ELECTRONIC CIRCUIT BOARDS, WIRING HARNESSES OR ELECTRONIC SUBASSEMBLIES WHILE THE UNIT IS UNDER POWER. 1. Inspect the received packages for external shipping damage. If damaged, please advise the shipper first, then Teledyne API.
Teledyne API – Model T700 Dynamic Dilution Calibrator Getting Started 3.2.1. FRONT PANEL Figure 3-1: T700 Front Panel Layout Figure 3-2: Display Screen and Touch Control The front panel liquid crystal display (LCD) screen includes touch control. Upon calibrator start-up, the LCD shows a splash screen and other initialization indicators before the main display appears, similar to Figure 3-2 above.
Getting Started Teledyne API – Model T700 Dynamic Dilution Calibrator Table 3-1: Display Screen and Touch Control Description Field Description/Function LEDs indicating the states of the calibrator: Definition Name Color State Unit is operating in STANDBY mode. Active Green This LED is lit when the instrument is actively producing calibration gas (GENERATE mode).
Getting Started Teledyne API – Model T700 Dynamic Dilution Calibrator 3.2.2. REAR PANEL Figure 3-4: T700 Rear Panel Layout Table 3-2 provides a description of each component on the rear panel. 06873B DCN6388...
Teledyne API – Model T700 Dynamic Dilution Calibrator Getting Started Table 3-2: Rear Panel Description Component Function Cools instrument by pulling ambient air into chassis through side vents and exhausting through rear. Connector for three-prong cord to apply AC power to the analyzer AC Power CAUTION! The cord’s power specifications (specs) MUST comply with the...
Motherboard CPU PCA Relay PCA DC Power supplies INPUT GAS Cal Gas Mass Flow Controller PRESSURE SENSOR PCA Diluent Mass Flow Controller ON / OFF Switch FRONT Figure 3-5: T700 Internal Layout – Top View – Base Unit 06873B DCN6388...
Switch Input Gas Optional 2nd Cal Gas Diluent FRONT Cal Gas Mass Flow Mass Flow Pressure Mass Flow Controller Controller Sensor PCA Controller Figure 3-6: T700 Internal Layout – Top View – with Optional O Generator and Photometer 06873B DCN6388...
ALWAYS WEAR GLASSES MADE FROM SAFETY UV FILTERING GLASS (PLASTIC GLASSES ARE INADEQUATE). Note The T700 calibrator is equipped with a universal power supply that allows it to accept any AC power configuration, within the limits specified in Table 2-2.
Getting Started 3.3.1.2. Connecting Analog Outputs The T700 is equipped with an analog output channel accessible through a connector on the back panel of the instrument. The standard configuration for this output is 0-5 VDC. It can be set by the user to output one of a variety of diagnostic test functions (see Section 4.10.1.2).
Getting Started Teledyne API – Model T700 Dynamic Dilution Calibrator STATUS Figure 3-8: Status Output Connector The pin assignments for the Status Outputs are: Table 3-3: Status Output Pin Assignments OUTPUT STATUS CONDITION DEFINITION SYSTEM OK On if no faults are present.
Teledyne API – Model T700 Dynamic Dilution Calibrator Getting Started 3.3.1.4. Connecting the Control Inputs The calibrator is equipped with 12 digital control inputs that can be used to initiate various user programmable calibration sequences (see Section 4.3.1.5 for instructions on assigning the control inputs to specific calibration sequences).
Getting Started Teledyne API – Model T700 Dynamic Dilution Calibrator There are two methods for energizing the control inputs. The internal +5V available from the pin labeled “+” is the most convenient method. However, if full isolation is required, an external 5 VDC power supply should be used.
3.3.1.5. Connecting the Control Outputs The calibrator is equipped with 12 opto-isolated, digital control outputs. These outputs are activated by the T700’s user-programmable calibration sequences (see Sections 4.3.1.6 and 4.3.2.8 for instructions on assigning the control inputs to specific calibration sequences).
8-watt valves based on the condition of the status block bits described below The option consists of a custom Printed Circuit Assembly (PCA) that mounts to the back of the T700 and a universal AC-to-DC power supply. Figure 3-11: T700 Rear Panel Valve Driver Installed...
Signal-I/O diagnostic function in the T700 software (see Section 9.4.11.5). However, the drive outputs are mapped in reverse to the status control bits such that Bit-0 (LSB) is valve drive 8 and Bit-7 is valve drive 1.
Page 50
For RS-232 communications with data terminal equipment (DTE) or with data communication equipment (DCE) connect either a DB9-female-to-DB9-female cable (Teledyne API part number WR000077) or a DB9-female-to-DB25-male cable (Option 60A), as applicable, from the analyzer’s rear panel RS-232 port to the device. Adjust the DCE-DTE switch (Figure 3-4) to select DTE or DCE as appropriate (Section 5.1).
Teledyne API – Model T700 Dynamic Dilution Calibrator Getting Started Figure 3-13: Rear Panel Connector Pin-Outs for RS-232 Mode The signals from these two connectors are routed from the motherboard via a wiring harness to two 10-pin connectors on the CPU card, J11 and J12 (Figure 3-14).
Figure 3-14: Default Pin Assignments for CPU COMM Port Connector (RS-232). Teledyne API offers two mating cables, one of which should be applicable for your use. P/N WR000077, a DB-9 female to DB-9 female cable, 6 feet long. Allows connection of the serial ports of most personal computers.
Page 53
Teledyne API – Model T700 Dynamic Dilution Calibrator Getting Started RS-232 COM PORT DEFAULT SETTINGS Received from the factory, the analyzer is set up to emulate a DCE or modem, with Pin 3 of the DB-9 connector designated for receiving data and Pin 2 designated for sending data.
Getting Started Teledyne API – Model T700 Dynamic Dilution Calibrator 3. Check that the following cable connections are made in all instruments (again refer to Figure 3-15): 4. J3 on the Multidrop/LVDS PCA to the CPU’s COM1 connector 5. (Note that the CPU’s COM2 connector is not used in Multidrop) 6.
Teledyne API – Model T700 Dynamic Dilution Calibrator Getting Started 11. On the rear panel of each analyzer, adjust the DCE DTE switch (see Figure 3-4 and Section 5.1) so that the green and the red LEDs (RX and TX) of the COM1 connector (labeled RS232) are both lit.
If the calibrator is also equipped with an internal photometer, the zero air source supply gas must be capable of a continuous rate of flow of at least 1.1 LPM. Zero Air can be purchased in pressurized canisters or created using a Teledyne API’s Model 701 Zero Air Generator.
Teledyne API – Model T700 Dynamic Dilution Calibrator Getting Started NIST TRACEABLE CALIBRATION GAS STANDARDS All calibration gases should be verified against standards of the National Institute for Standards and Technology (NIST). To ensure NIST traceability, we recommend acquiring cylinders of working gas that are certified to be traceable to NIST Standard Reference Materials (SRM).
Getting Started Teledyne API – Model T700 Dynamic Dilution Calibrator Table 3-7: NIST Standards for CO Nominal Amount of Description Substance 2612a Carbon Monoxide in Air 10 ppm 2613a Carbon Monoxide in Air 20 ppm 2614a Carbon Monoxide in Air...
Teledyne API – Model T700 Dynamic Dilution Calibrator Getting Started Table 3-11: NIST Standards for SO Nominal Amount of Description substance 1661a Sulfur Dioxide in Nitrogen 1662a Sulfur Dioxide in Nitrogen 1000 ppm 1663a Sulfur Dioxide in Nitrogen 1500 ppm...
7. 4SLPM x 1.5 = 6SLPM = Total Gas Flow Rate 8. If the T700 calibrator so that the cal gas flow rate is 2SLPM (therefore the Diluent Flow Rate would need to be set at 4 SLPM) the Calibration Gas ratio would be: 9.
Figure 3-17: Set up for T700 – Connecting the Basic T700 to a Sample Manifold To determine if the gas flow on the vent line is ≥ 5 SLPM subtract the gas flow for each instrument connected to the outlets of the T700 from the TOTAL FLOW setting for the calibrator (see Section 3.4.9).
Teledyne API – Model T700 Dynamic Dilution Calibrator CONNECTING THE CALIBRATOR TO A SAMPLE GAS MANIFOLD Use this setup when connecting the T700 calibrator to an analyzer network using a sample manifold. In this case, the sampling cane and the manifold itself act as the vent for the T700.
(such as an analyzer). Figure 3-19: Set up for T700 – Connecting the T700 to a Calibration Manifold • This method requires the analyzers connected to the calibration system have Note internal zero/span valves.
Another type of calibration setup utilizes separate span gas and the zero air manifolds (see Figure 3-20). Figure 3-20: Set up for T700 – Connecting the T700 to a Dual Span Gas / Zero Air Manifold This set up is subject to the same notes and conditions as the single calibration...
Figure 3-21: T700 Pneumatic Diagram – Base Unit The standard T700 Dynamic Dilution Calibrator is equipped with one calibration gas mass flow controller (flow rate and one diluent gas mass flow controller 0 –...
Getting Started Teledyne API – Model T700 Dynamic Dilution Calibrator the O generator, allow the T700 to use the GPT technique to more precisely create NO calibration mixtures On Back Panel Figure 3-22: Internal Pneumatics for T700 Calibrator with Optional O Generator and GPT Chamber.
Page 67
The voltage output of a reference detector, also internal to the generator, is digitized and sent to the T700’s CPU where it is used as input for a control loop that maintains the intensity of the UV lamp at a level appropriate for the chosen set point.
In an instrument with multiple MFCs the CPU chooses which MFC to use depending on the target gas flow requested. When generating O or in GPT Pre-Set mode, the photometer pump is the primary creator of gas flow through the T700. Flow rates are controlled by critical flow orifice(s) located in the gas stream...
Page 69
An optional third mass flow controller can be added on the calibration gas stream. When this option is installed the T700 has both calibration gas MFCs on the same gas stream, installed in parallel (see Figure 3-24 and Figure 3-25). The calibrator turns on the MFC with the lowest flow rate that can accommodate the requested flow and can therefore supply the most accurate flow control.
Getting Started Teledyne API – Model T700 Dynamic Dilution Calibrator On Back Panel Figure 3-24: Basic T700 with Multiple Calibration Gas MFCs 06873B DCN6388...
If the tube is kept at constant temperature, usually about 50C, the device will provide a stable source of gas for a year or more. A pneumatic diagram of the T700 with this option is shown in Figure 3-27, including the generator.
Teledyne API – Model T700 Dynamic Dilution Calibrator Getting Started Figure 3-27: Pneumatic Diagram of T700 with Permeation Generator 06873B DCN6388...
Page 74
Getting Started Teledyne API – Model T700 Dynamic Dilution Calibrator This page intentionally left blank. 06873B DCN6388...
Page 75
Once the T700 has stabilized, the response to the permeation tube is not expected to change more than 5% if the zero air is provided for Teledyne API’s M701 or other dry zero air source.
Getting Started Teledyne API – Model T700 Dynamic Dilution Calibrator 3.3.3. PERMEATION TUBE SETUP FOR THE T700 1. Press SETUP and GAS 2. Press PERM 3. Enter the elution rate for the permeation tube and select the type of gas by pressing the gas button to scroll through the gas list until the desired gas is shown.
Teledyne API – Model T700 Dynamic Dilution Calibrator Getting Started 3.3.4. PERMEATION TUBE CALCULATION The permeation tube concentration is determined by the permeation tube’s specific output or elution rate (which is normally stated in ng/min), the permeation tube temperature (C) and the air flow across it (slpm). The elution rate of the tube is normally stated at an operating temperature of 50°C and is usually printed on the tube's...
Use the following menu sequence: GEN>AUTO>[Select A or B]>ENTR. Your chosen output is now selected for calibration. The following illustration depicts the pneumatic flow for the T700 Calibrator with the optional dual gas output at the output valve. 06873B DCN6388...
– Special) 3.4. STARTUP, FUNCTIONAL CHECKS, AND INITIAL CALIBRATION If you are unfamiliar with the T700 principles of operation, we recommend that you read Section 10. For information on navigating the calibrator’s software menus, see the menu trees described in Appendix A.
Getting Started Teledyne API – Model T700 Dynamic Dilution Calibrator conditions for 30 minutes after power up. If warning messages persist after the 30 minutes warm up period is over, investigate their cause using the troubleshooting guidelines in Section 9 of this manual.
Page 81
Teledyne API – Model T700 Dynamic Dilution Calibrator Getting Started MESSAGE MEANING PHOTO REFERENCE WARNING The photometer reference reading is outside of allowable limits. The calibrator’s motherboard was not detected during power up. - THIS WARNING only appears on Serial I/O COMM Port(s).
Getting Started Teledyne API – Model T700 Dynamic Dilution Calibrator 3.4.3. FUNCTIONAL CHECKS 5. After the calibrator’s components have warmed up for at least 30 minutes, verify that the software properly supports any hardware options that are installed. 6. Check to ensure that the calibrator is functioning within allowable operating parameters.
It is not recommended that ammonia be used in the T700. 3.4.6. USER DEFINED GAS TYPES 3.4.6.1. User Defined Gas Types – General The T700 calibrator can accept up to four different user defined gases. This allows the use of: ...
Page 84
Alternatively, both bottles can be assigned user defined names; e.g. CO2A and CO2B User defined gas names are added to the T700’s gas library and will appear as choices during the various calibrator operations along with the default gas names listed in Section 3.4.5.
Teledyne API – Model T700 Dynamic Dilution Calibrator Getting Started 3.4.6.2. User Defined Gas Types – Defining the Gas Name In this example, we will be using PROPANE (C ). Press: STANDBY A-CAL=0.000 LPM Make sure that the T700 <TST TST>...
Getting Started Teledyne API – Model T700 Dynamic Dilution Calibrator 3.4.6.3. User Defined Gas Types – Setting the MOLAR MASS The molar mass of a substance is the mass, expressed in grams, of 1 mole of that specific substance. Conversely, one mole is the amount of the substance needed for the molar mass to be the same number in grams as the atomic mass of that substance.
Page 87
Teledyne API – Model T700 Dynamic Dilution Calibrator Getting Started To set the molar mass of a user defined gas, press: Note If the contents of the bottle are predominantly N , use the molar mass of N (28.01). 06873B DCN6388...
ENABLED. Any of these can be disabled. Disabling a gas type means that it does not appear in certain prompts during portions of the T700’s operation (e.g. setting up sequences) and is not figured into the calibrators calculating when determining calibration mixtures.
Teledyne API – Model T700 Dynamic Dilution Calibrator Getting Started 3.4.7. DEFINING CALIBRATION SOURCE GAS CYLINDERS 3.4.7.1. Setting Up the Ports with Single Gas Cylinders To program the T700 calibrator’s source gas input ports for a single gas cylinder, press: 06873B DCN6388...
0.000 parts per thousand 00.00 Repeat the above steps for each of the T700 calibrator’s four gas inlet ports. If no gas is present on a particular port, leave it set at the default setting of NONE. 06873B DCN6388...
Teledyne API – Model T700 Dynamic Dilution Calibrator Getting Started 3.4.7.2. Setting Up the Ports with Multiple Gas Cylinders When an application utilizes multiple gas cylinders, program as follows (note that the GENERATE>AUTO menu (Section 4.2.6) differs from that for a single gas (Section 4.2.1):...
This setting will be the default mode (if not equipped with the photometer option) of the T700 calibrator and will be used whenever the calibrator is using the GENERATE AUTO command or the GENERATE sequence step to create a calibration mixture.
Getting Started 3.4.9. SETTING THE T700’S TOTAL GAS FLOW RATE The default total gas flow rate for the T700 Dynamic Dilution Calibrator is 2 LPM. The calibrator uses this flow rate, along with the concentrations programmed into the calibrator for the component gas cylinders during set up, to compute individual flow rates for both diluent gas and calibration source gases in order to produce calibration mixtures that match the desired output concentrations.
Page 94
The GENERATE MANUAL menu (see Section 4.2.2) or; As part of a MANUAL step when programming a sequence (see Section 4.3.2.9). Note When calculating total required flow for T700’s with O photometers installed ensure to account for the 800 cc/min flow it requires. 06873B DCN6388...
25:00:00. Once you adjust the setting to an allowable value, the ENTR button will reappear. The T700 calibrator software has a variety of operating modes, which are controlled from the front panel touch screen. (Remote operation is described in Section 6). The most common mode that the calibrator will be operating in is the STANDBY mode.
Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator Figure 4-1: Front Panel Display The mode field of the front panel display indicates to the user which operating mode the unit is currently running.
This mode is not available in units without O generators installed. This mode is not available in units without internal photometers installed. The revision of the Teledyne API software installed in this calibrator will be displayed following the word SETUP. E.g. “SETUP G.4” 06873B DCN6388...
Teledyne API – Model T700 Dynamic Dilution Calibrator 4.1. STANDBY MODE When the T700 Dynamic Dilution Calibrator is in standby mode, it is at rest. All internal valves are closed except the diluent inlet valve. The mass flow controllers are turned off.
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation On Back Panel Figure 4-2: Gas Flow through T700 with O Generator and Photometer Options during STANDBY 4.1.1. TEST FUNCTIONS A variety of TEST functions are available for viewing at the front panel whenever the calibrator is in STANDBY Mode.
Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator To view these TEST functions, press: Figure 4-3: Viewing T700 Test Functions 06873B DCN6388...
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation Table 4-3: Test Functions Defined DISPLAY UNITS DESCRIPTION The actual gas flow rate of source gas being output by the calibrator. A-CAL T-CAL Target source gas flow rate for which the calibrator output is set.
Mass Flow Controller Figure 4-4: Gas Flow through Basic T700 in GENERATE Mode Table 4-4 shows the status of the T700’s various pneumatic components when the calibrator is in GENERATE mode: Table 4-4: Status of Internal Pneumatics During GENERATE Mode...
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation Figure 4-5: Gas Flow through T700 with O Options when Generating Non-O Source Gas Figure 4-6: Gas Flow through T700 with O Options when Generating O...
The target concentration, and; The TOTAL FLOW to be output by the T700. Using this information, the T700 calibrator automatically calculates and sets the individual flow rates for the Diluent and chosen component gases to create the desired calibration mixture.
Page 107
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation To use the GENERATE AUTO feature, press: 06873B DCN6388...
• The minimum total flow should equal 150% of the flow requirements of all of the instruments to which the T700 will be supplying calibration gas. • Example: If the T700 is will be expected to supply calibration gas mixtures simultaneously to a system in composed of three analyzers each requiring 2 LPM , the proper Total Flow output should be set at: (2 + 2 + 2) x 1.5 = 9.000 LPM...
The limits are <10% of the lowest rated MFC or >100% of the combined full-scale ratings for both mass flow controllers. • The T700 will automatically select the MFC with the lowest flow rate that can accommodate the requested flow, thereby affording the most precise flow control.
Menu In the following demonstration we will be using the values from the examples given with Equations 6-1 and 6-2 above and assume a T700 calibrator with at least one source gas mass flow controller capable of 3.0 LPM output.
0.105 LPM) and the GPT chamber’s volume is known, once the TOTAL GAS FLOW requirements, the source concentration of NO, and the target concentration for the O generator are entered into the calibrator’s software. The T700 adjusts the NO flow rate and diluent (zero air) flow rate to create the appropriate NO concentration at the output.
The limits are <10% of the lowest rated MFC or >100% of the combined full-scale ratings for both mass flow controllers. • The T700 will automatically select the MFC with the lowest flow rate that can accommodate the requested flow, thereby affording the most precise flow control.
Overview of Operating Modes and Basic Operation 4.2.3.4. T700 Calibrator GPT Operation The following table and figures show the status of the T700’s internal pneumatic components and internal gas flow when the instrument is in GPT generating modes. Table 4-5: Status of Internal Pneumatics During GENERATE GPT Mode...
Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator 4.2.3.5. Initiating a GPT Calibration Gas Generation Note It is highly recommended to perform a GPT Pre-Set before initiating any GPT gas generation (Section 4.2.4).
NO gas that would normally be used. 4.2.4.1. T700 Calibrator GPTPS Operation The following table and figures show the status of the T700’s internal pneumatic components and internal gas flow when the instrument is in GPTPS generating modes.
Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator Figure 4-8: Gas Flow through T700 with O Options when in GPTPS Mode 06873B DCN6388...
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation 4.2.4.2. Initiating a GPT Pre-Set To activate the GPTPS feature you will need to know: The TOTAL GAS FLOW for the mixture output; ...
Teledyne API – Model T700 Dynamic Dilution Calibrator 4.2.5. GENERATE PURGE: ACTIVATING THE T700’S PURGE FEATURE The T700 calibrator’s PURGE feature clears residual source gases and calibration mixtures gases from the previous generated steps from the instruments internal pneumatics as well as any external pneumatic lines down stream from the calibrator.
Page 119
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation To activate the PURGE feature, press: MPORTANT MPACT ON EADINGS OR This PURGE feature does not stop automatically. Manually press the STBY button to stop the purging process..
NO or CH4) using the GEN AUTO, GEN MANUAL buttons or a preprogrammed calibration SEQUENCE, press: Note The ACT> button only appears if the T700 is generating gas from a multiple-gas cylinder. To start any preprogrammed calibration SEQuences, first place the calibrator in STANDBY mode (the SEQ button replaces the ACT> button)
(zero) causes the sequence to execute indefinitely. CC INPUT Specifies which of the T700’s Digital Control Inputs will initiate the sequence. Specifies which of the T700’s Digital Control Outputs will be set when the sequence CC OUTPUT is active. Enables or disables an internal automatic timer that can initiate sequences using the TIMER ENABLE T700’s built in clock.
Up to 5 levels of nested sequences can be programmed. Allows the sequence to activate the T700’s digital control outputs. Similar to the CC SETCCOUTPUT OUPUT attribute, but can be set and reset by individual steps.
To create a sequence, use the instructions in the following sections to name the sequence, set its associated parameters and define the steps to be included. 4.3.1.1. Activating a Sequence from the T700 Front Panel To activate an already programmed sequence from the front panel, press:...
Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator 4.3.1.2. Naming a Sequence The first step of creating a calibration sequence is to assign it a name. The name can be up to 10 characters and can be comprised of any alpha character (A to Z), and numeral (0 to 9) or the underscore character (“_“).
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation 4.3.1.3. Setting the Repeat Count for a Sequence The sequence can be set to repeat a certain number of times, from 1 to 100. It can also be set to repeat indefinitely by inputting a zero (0) into the REPEAT COUNTER.
4.3.1.4. Using the T700’s Internal Clock to Trigger Sequences Sequences can be set to trigger based on the T700’s internal clock. The sequence can be set up to start at a predetermined date and time. It can also be set to repeat after a predetermined delay time.
Page 127
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation To specify a starting date and time for the sequence, press: STANDBY A-CAL=0.000 LPM Make sure that the T700 is <TST TST> GEN STBY SEQ SETUP in standby mode.
Page 128
Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator To set the delta timer, press: STANDBY A-CAL=0.000 LPM Make sure that the T700 is <TST TST> GEN STBY SEQ SETUP in standby mode. SETUP X.X...
This feature allows the calibrator to operate in a slave mode so that external control sources, such as a data logger can initiate the calibration sequences. Each of the T700 calibrator’s control outputs is located on the back of the instrument (see Figure 3-4).
Teledyne API – Model T700 Dynamic Dilution Calibrator 4.3.1.6. Setting Up Control Outputs for a Sequence The T700 calibrator’s control outputs allow the calibrator to control devices that accept logic-level digital inputs, such as programmable logic controllers (PLCs), data loggers, or digital relays/valve drivers.
Page 131
Toggle this button to turn the selected bit ON/OFF (0 or 1). Each bit shown on the display represents one of the control output pins located on the back of the T700 (see Figure 3-2), The left most bit is Bit 1, the next bit to the right, bit 2,...
Teledyne API – Model T700 Dynamic Dilution Calibrator 4.3.1.7. Setting the PROGRESS Reporting Mode for the Sequences As sequences run, the T700 calibrator reports progress by displaying a message in the MODE field of the front panel display (See Figure 3-1). There are several types of report modes available (see Table 4-10).
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation 4.3.2. ADDING SEQUENCE STEPS To insert an instruction step into a sequence, navigate to the INSERT STEP submenu by pressing: STANDBY A-CAL=0.000 LPM Make sure that the T700 is <TST...
4.3.2.1. The GENERATE Step This step operates and is programmed similarly to the GENERATE AUTO. At the end of the programming sequence, the T700 firmware will automatically insert a DURATION step that needs to be defined. To insert a GENERATE step into a sequence, press:...
This step operates and is programmed similarly to the GENERATE GPT (see Section 4.2.3 for information on choosing the correct input values for this step). At the end of the programming sequence, the T700 firmware will automatically insert a DURATION step that needs to be defined.
This step operates and is programmed similarly to the GENERATE GPTPS (see Section 4.2.4 for information on choosing the correct input values for this step). At the end of the programming sequence, the T700 firmware will automatically insert a DURATION step that needs to be defined.
4.3.2.4. The PURGE Step This step places the T700 into PURGE mode. At the end of the programming sequence, the T700 firmware will automatically insert a DURATION step that needs to be defined. To insert a PURGE step into a sequence, press: 4.3.2.5.
Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator 4.3.2.6. The DURATION Step The duration step causes the T700 to continue performing whatever action was called for by the preceding step of the sequence. ...
Page 139
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation To insert an EXECSEQ step into a sequence, press: 06873B DCN6388...
Teledyne API – Model T700 Dynamic Dilution Calibrator 4.3.2.8. The CC OUTPUT Step This instruction causes the sequence to set or reset the T700’s digital control outputs. It is very useful in situations where the control outputs are being used to trigger other devices that need to be turned off and on in synch with the operation of the calibrator as it progress through the sequence.
The MANUAL step causes the T700 calibrator to enter MANUAL CALIBRATION MODE. It is programmed in a similar manner to the calibrator’s GENERATE MANUAL function. AT the end of the programming sequence, the T700 firmware will automatically insert a DURATION step that needs to be defined.
Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator 4.3.2.10. Deleting or Editing an Individual Step in a Sequence To delete or edit an individual step in an existing Sequence, press: 06873B DCN6388...
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation 4.3.3. DELETING A SEQUENCE To delete a sequence from the T700 calibrator’s memory, press: STANDBY ACT CAL=0.000 LPM Make sure that the T700 is <TST TST>...
Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator 4.4. SETUP CFG Pressing the CFG button displays the instrument’s configuration information. This display lists the calibrator model, serial number, firmware revision, software library revision, CPU type and other information.
CLOCK AND ADJUSTING SPEED 4.5.1. SETTING THE INTERNAL CLOCK’S TIME AND DAY The T700 has a time of day clock that supports the DURATION step of the calibration sequence feature, time of day TEST function, and time stamps on most COMM port messages.
Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator 4.5.2. ADJUSTING THE INTERNAL CLOCK’S SPEED In order to compensate for CPU clocks which run faster or slower, you can adjust a variable called CLOCK_ADJ to speed up or slow down the clock by a fixed amount every day.
Overview of Operating Modes and Basic Operation 4.6. SETUP PASS The T700 provides password protection of the calibration and setup functions to prevent unauthorized adjustments. When the passwords have been enabled in the PASS menu item, the system will prompt the user for a password anytime a password-protected function is requested.
Page 148
Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator Example: If all passwords are enabled, the following touch screen button sequence would be required to enter the VARS or DIAG submenus: STANDBY A-CAL=0.000 LPM <TST...
Each type of Teledyne API’s calibrator is configured with a default ID code. The default ID code for all T700 calibrators is typically 700 (or 0). The ID number is only important if more than one calibrator is connected to the same communications channel such as when several calibrators are on the same Ethernet LAN (See Section 5.4);...
VARS menu. The following table lists all variables that are available within the 818 password protected level. See Appendix A2 for a detailed listing of all of the T700 variables that are accessible through the remote interface.
Page 151
- sign if the clock is too fast (See Default=0 Section 4.5). DO NOT ADJUST OR CHANGE these values unless instructed to by Teledyne API’s customer service personnel. Only available in calibrators with O photometer and generator options installed. MPORTANT...
Page 152
Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator To access and navigate the VARS menu, use the following button sequence: STANDBY A-CAL=0.0000 LPM Make sure that the T700 is in standby mode. <TST TST>...
4.10.1. TEST CHAN OUTPUT: USING THE TEST CHANNEL ANALOG OUTPUT The T700 calibrator comes equipped with one analog output. It can be set by the user to carry the current signal level of any one of the parameters listed in Table 4-14 and will output an analog VDC signal that rises and falls in relationship with the value of the parameter.
Page 154
Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator To configure the calibrator’s TEST CHANNEL, set the electronic signal type of each channel and calibrate the outputs. This consists of: Choosing a Test Channel function to be output on the channel (Table 4-14).
Page 155
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation To access the analog I/O configuration sub menu, press: 06873B DCN6388...
Teledyne API – Model T700 Dynamic Dilution Calibrator 4.10.1.2. Selecting a Test Channel Function to Output The Test Functions available to be reported are listed on Table 4-14: Table 4-14: Test Channels Functions available on the T700’s Analog Output TEST CHANNEL DESCRIPTION...
Page 157
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation To activate the TEST Channel and select a function press: STANDBY A-CAL=0.0000 LPM <TST TST> GEN STBY SEQ SETUP Make sure that the T700 is in standby mode.
Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator 4.10.1.3. Test Channel Voltage Range Configuration In its standard configuration, the analog outputs is set to output a 0 – 5 VDC signals. Several other output ranges are available (see Table 4-15). Each range is usable from - 5% to + 5% of the rated span.
Overview of Operating Modes and Basic Operation 4.10.1.4. Turning the Test Channel Over-Range Feature ON/OFF In its default configuration, a ± 5% over-range is available on each of the T700’s TEST CHANNEL output. This over-range can be disabled if your recording device is sensitive to excess voltage or current.
Some analog signal recorders require that the zero signal is significantly different from the baseline of the recorder in order to record slightly negative readings from noise around the zero point. This can be achieved in the T700 by defining a zero offset, a small voltage (e.g., 10% of span).
Page 161
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation 06873B DCN6388...
Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator 4.10.1.6. Test Channel Calibration TEST CHANNEL calibration needs to be carried out on first startup of the calibrator (performed in the factory as part of the configuration process) or whenever recalibration is required.
Page 163
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation AUTOMATIC TEST CHANNEL CALIBRATION Note Before performing this procedure, ensure that the AUTO CAL feature is turned OFF for MFC_DRIVE_1, MFC_DRIVE_2 and MFC_DRIVE_3 if installed). Ensure that the AUTO CAL feature is turned ON for the TEST CHANNEL (see Section 4.10.1.6).
Page 164
Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator To initiate an automatic calibration from inside the TEST CHANNEL submenu, press: 06873B DCN6388...
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation MANUAL CALIBRATION OF THE TEST CHANNEL CONFIGURED FOR VOLTAGE RANGES For highest accuracy, the voltages of the analog outputs can be calibrated manually. Note The menu for manually adjusting the analog output signal level will only appear if the AUTO-CAL feature is turned off for the channel being adjusted (see Section 4.10.1.6).
Page 166
Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator To adjust the signal levels of an analog output channel manually, press: From the AIO CONFIGURATION SUBMENU (See Section 6.9.1.1.) DIAG ANALOG I/O CONFIGURATION PREV NEXT...
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation 4.10.1.7. AIN Calibration This is the sub-menu in which to calibrate the A-to-D conversion circuitry (Sections 9.4.11.1 and 10.3.5.1). This calibration is only necessary after a major repair such as the replacement of a CPU, a motherboard or a power supply.
OPERATING LEVELS 4.11.1. GENERAL INFORMATION ABOUT LEADS LEVELS The T700 calibrator can be equipped with a version of firmware that includes support for LEADS, a data collection and analysis system LEADS specifically designed for handling meteorological and environmental data particularly when there is a need to integrate data and control instrumentation from several different manufacturers.
Configuration for one or both of two status output blocks. Up to twenty levels can be defined and used with the T700 using a range of ID numbers from 0-98. Level 99 is reserved for standby. The levels are not time based and do not include characteristics such as start time or duration, therefore a single LEVEL can not switch between different concentration levels and flow rates.
Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator 4.11.5. PROGRAMMING NEW LEVELS To begin programming a new LEVEL find the LVL submenu by pressing: STANDBY A-CAL=0.000 LPM Make sure that the T700 is <TST...
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation 4.11.5.1. Creating a GENERATE LEVEL To create a LEVEL using the T700’s AUTO generation function, press: Starting at the CHOOSE ACTION Submenu CHOOSE ACTION Submenu SETUP X.X...
Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator 4.11.5.2. Creating a GPT LEVEL To create a LEVEL using the T700’s GPT function, press: Starting at the CHOOSE ACTION Submenu CHOOSE ACTION Submenu SETUP X.X...
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation 4.11.5.3. Creating a GPTPS LEVEL To create a LEVEL using the T700’s GPTPS function, press: Starting at the CHOOSE ACTION Submenu CHOOSE ACTION Submenu SETUP X.X...
Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator 4.11.5.4. Creating a MANUAL LEVEL To create a level using the T700’s MANUAL generation function, press: Starting at the CHOOSE ACTION Submenu CHOOSE ACTION Submenu SETUP X.X...
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation 4.11.5.5. Editing or Deleting a LEVEL To edit or delete an existing LEVEL, press: STANDBY A-CAL=0.000 LPM Make sure that the T700 is in standby mode.
BLOCK 1: This block corresponds to the physical CONTROL OUTPUT connections located on the back panel of the T700 (see Figure 3-4 and Section 3.3.1.5). BLOCK 2: The second status block does not correspond to any physical output but is used to communicate status over the serial data port.
Receive and Transmit lines on RS-232 emulating a cross-over or null- modem cable. The switch has no effect on COM2. The T700 is equipped with two serial communication ports (labeled RS232 and COM2), a USB com port and an Ethernet port located on the rear panel. The two serial ports are accessible via two DB-9 connectors (see Figure 3-4): RS232 (COM1), a male DB-9 connector, and COM2, a female DB9 connector.
Communications Setup and Operation Teledyne API – Model T700 Dynamic Dilution Calibrator capable of operating at 10BaseT or 100BaseT. DHCP is enabled by default (Section 5.4.1). This configuration is useful for quickly getting an instrument up and running on a network.
Page 179
Modes are listed in the order in which they appear in the SETUP MORE COMM COM[1 OR 2] MODE menu The default setting for this feature is ON. Do not disable unless instructed to by Teledyne API’s Customer Service personnel.
Page 180
Communications Setup and Operation Teledyne API – Model T700 Dynamic Dilution Calibrator 06873B DCN6388...
Teledyne API – Model T700 Dynamic Dilution Calibrator Communications Setup and Operation 5.2.2. COM PORT BAUD RATE To select the baud rate of either COM Port, go to SETUP>MORE>COMM and select either COM1 or COM2 as follows (use COM2 to view/match your personal computer...
Communications Setup and Operation Teledyne API – Model T700 Dynamic Dilution Calibrator 5.2.3. COM PORT TESTING The serial ports can be tested for correct connection and output in the COMM menu. This test sends a string of 256 ‘w’ characters to the selected COM port. While the test is running, the red LED on the rear panel of the calibrator should flicker.
Flickers during any activity on the LAN. 5.4.1. CONFIGURING THE ETHERNET INTERFACE USING DHCP The Ethernet feature for your T700 uses Dynamic Host Configuration Protocol (DHCP) to configure its interface with your LAN automatically. This requires your network servers also be running DHCP. The calibrator will do this the first time you turn the instrument on after it has been physically connected to your network.
T700 for all Teledyne API’s T700 calibrators is “T700”, the host name may be changed to fit customer needs. Do not change the setting for this property unless instructed to by Teledyne API’s Customer Service personnel. Note If the gateway IP, instrument IP and the subnet mask are all zeroes (e.g.
Page 185
Teledyne API – Model T700 Dynamic Dilution Calibrator Communications Setup and Operation To view the above properties listed in Table 5-3, press: 06873B DCN6388...
Communications Setup and Operation Teledyne API – Model T700 Dynamic Dilution Calibrator 5.4.1.1. Manually Configuring the Network IP Addresses There are several circumstances when you may need to manually set the Ethernet configuration: Your LAN is not running a DHCP software package, ...
Page 187
Teledyne API – Model T700 Dynamic Dilution Calibrator Communications Setup and Operation Internet Configuration Touchscreen Button Functions From Step 1 above) BUTTON FUNCTION Press to cycle through the range of numerals and available characters (“0 – 9” & “ . ”) <CH CH>...
The HOSTNAME is the name by which the calibrator appears on your network. The default name for all Teledyne API’s T700 calibrators is T700. To change this name (particularly if you have more than one T700 calibrator on your network), press.
The analyzer can be operated through a personal computer by downloading the TAPI USB driver and directly connecting their respective USB ports. 1. Install the Teledyne T-Series USB driver on your computer, downloadable from the Teledyne API website under Help Center>Software Downloads (www.teledyne- api.com/software).
Page 190
Communications Setup and Operation Teledyne API – Model T700 Dynamic Dilution Calibrator 5. In the instrument’s SETUP>MORE>COMM>COM2 menu, make the following settings: Baud Rate: 115200 COM2 Mode Settings: Quiet Mode Computer Mode MODBUS RTU MODBUS ASCII E,8,1 MODE E,7,1 MODE...
It assumes that the electrical connections have been made as described in Section3.3.1 The T700 can be remotely configured, calibrated or queried for stored data th through the rear panel serial ports, via either Computer mode (using a personal computer with a dedicated interface program) or Interactive mode (using a terminal emulation program).
Remote Operation Teledyne API – Model T700 Dynamic Dilution Calibrator Figure 6-1: APICOM Remote Control Program Interface 6.2. INTERACTIVE MODE Interactive mode is used with a terminal emulation programs or a “dumb” computer terminal. 6.2.1. REMOTE CONTROL VIA A TERMINAL EMULATION PROGRAM Start a terminal emulation program such as HyperTerminal.
Teledyne API – Model T700 Dynamic Dilution Calibrator Remote Operation COMMAND Function be sent to the calibrator to be executed until this is done. On personal computers, this is achieved by pressing the ENTER button. Erases one character to the left of the (backspace) cursor location.
Remote Operation Teledyne API – Model T700 Dynamic Dilution Calibrator Table 6-2: Teledyne API Serial I/O Command Types COMMAND COMMAND TYPE Calibration Diagnostic Logon Test measurement Variable Warning 6.2.1.3. Data Types Data types consist of integers, hexadecimal integers, floating-point numbers, Boolean expressions and text strings.
The DTE-DCE is in the DCE position. The T700 COMM port is set for a baud rate that is compatible with the modem. The Modem is designed to operate with an 8-bit word length with one stop bit.
Page 196
Remote Operation Teledyne API – Model T700 Dynamic Dilution Calibrator To change this setting, press: STANDBY A-CAL=0.000 LPM <TST TST> GEN STBY SEQ SETUP SETUP X.X PRIMARY SETUP MENU SETUP X.X COM1 MODE:0 SEQ CFG CLK PASS MORE EXIT <SET SET>...
Page 197
Teledyne API – Model T700 Dynamic Dilution Calibrator Remote Operation To initialize the modem, press: 06873B DCN6388...
6.4. PASSWORD SECURITY FOR SERIAL REMOTE COMMUNICATIONS In order to provide security for remote access of the T700, a LOGON feature can be enabled to require a password before the instrument will accept commands. This is done by turning on the SECURITY MODE (Mode 4, Section 5.2.1). Once the SECURITY MODE is enabled, the following items apply.
7.1. VIEWING THE PERFORMANCE STATISTICS FOR THE T700’S MFC’S It is possible to view the target flow rate, actual flow rate and actual gas pressure for each MFC via the FLOW submenu in the T700 calibrator (in real time). To access this information, press: 06873B DCN6388...
Page 200
Calibration and Verification Teledyne API – Model T700 Dynamic Dilution Calibrator 06873B DCN6388...
7.2. CALIBRATING THE OUTPUT OF THE T700’S MFC’S A table exists in the memory of the T700’s for each MFC that sets the output of the MFC at each of 20 equally spaced control points along its entire performance range.
Note A separate flow meter is required for the procedure. 1. Turn off the T700 Dynamic Dilution Calibrator. 2. Open the front panel to the T700 calibrator. This is the easiest access to the MFC output ports. A locking screw located at the top center of the front panel (See Figure 3-1) must be removed before the panel can be opened.
Teledyne API – Model T700 Dynamic Dilution Calibrator Calibration and Verification 7.2.2. VERIFYING AND CALIBRATING THE T700’S MFC’S Once the external flow meter is connected to the output of the MFC being verified/calibrated, perform the following steps: 06873B DCN6388...
For calibrators equipped with the O photometer, the accuracy of calibration mixtures involving O produced by the T700 depends entirely on the accuracy of the photometer, therefore it is very important that the photometer is operating properly and accurately. Setup for Verifying O Photometer Performance is shown in Section 7.3.1.
Teledyne API – Model T700 Dynamic Dilution Calibrator Calibration and Verification 7.3.2. VERIFYING O3 PHOTOMETER PERFORMANCE To verify the performance of the T700’s optional internal photometer perform the following steps: STANDBY A-CAL=0.0000 LPM Make sure that the T700 is in STANDBY <TST...
7.3.3.1. Setup Using Direct Connections Figure 7-3 shows the external zero air and O sources as well as the reference photometer connected directly to the fixtures on the back of the T700 Calibrator. Figure 7-3: External Photometer Validation Setup – Direct Connections Note A Minimum of 1.1 LPM is required for the external zero air source.
7.3.3.2. Setup Using a Calibration Manifold Figure 7-4 shows the external zero air and O sources as well as the reference photometer connected to the T700 Calibrator via calibration manifolds for both zero air and O Figure 7-4: External Photometer Validation Setup with Calibration Manifolds...
Calibration and Verification Teledyne API – Model T700 Dynamic Dilution Calibrator 7.3.4.1. Photometer Zero Calibration To set the zero point offset for the T700 Dynamic Dilution Calibrator’s photometer, press: 06873B DCN6388...
Teledyne API – Model T700 Dynamic Dilution Calibrator Calibration and Verification 7.3.4.2. Photometer Span Calibration To set the response SLOPE for the T700 Dynamic Dilution Calibrator’s photometer, press: 06873B DCN6388...
Calibration and Verification Teledyne API – Model T700 Dynamic Dilution Calibrator 7.3.5. O3 PHOTOMETER DARK CALIBRATION The Dark Calibration Test turns off the Photometer UV Lamp and records any offset signal level of the UV Detector-Preamp-Voltage to Frequency Converter circuitry. This...
To calibrate the flow of gas through the T700 calibrator’s optional photometer bench. 1. Turn OFF the T700 Dynamic Dilution Calibrator. 2. Attach the flow meter directly to the EXHAUST port of the T700 calibrator. 3. Turn the T700 Dynamic Dilution Calibrator ON.
Calibration and Verification Teledyne API – Model T700 Dynamic Dilution Calibrator 7.3.7. O3 PHOTOMETER BACKPRESSURE COMPENSATION CALIBRATION Any time there is a pneumatic configuration change, there is risk of impacting the internal measure/reference pressure. To compensate for this, a backpressure compensation calibration is required each time.
Teledyne API – Model T700 Dynamic Dilution Calibrator Calibration and Verification 7.4. CALIBRATING THE O GENERATOR 7.4.1. SETUP FOR VERIFICATION AND CALIBRATION THE O GENERATOR Note An external reference photometer is required for the procedure. 7.4.1.1. Setup Using Direct Connections Figure 7-5 shows the reference photometer connected directly to the fixtures on the back of the T700 Calibrator.
Page 214
(Default = 2.000 LPM) GENERATE ACT CAL=2.000 LPM <SET SET> GEN STBY SEQ SETUP Record O concentration from reference photometer Note The readings recorded from the T700’s A-CAL test function and the reference photometer should be within 1% of each other. 06873B DCN6388...
The T700 calibrator’s software includes a routine for automatically calibration the O generator. A table of drive voltages stored in the T700’s memory is the basis for this calibration. For each point included in the table used by the T700 to calibrate the optional O generator the user can set a drive voltage and a dwell time for that point.
Calibration and Verification Teledyne API – Model T700 Dynamic Dilution Calibrator 7.4.2.2. Adding or Editing O Generator Calibration Points To add a calibration point to the table or edit an existing point, press: STANDBY A-CAL=0.000 LPM Make sure that the T700 is in standby mode.
Teledyne API – Model T700 Dynamic Dilution Calibrator Calibration and Verification 7.4.2.3. Deleting O Generator Calibration Points To delete an existing calibration point, press: STANDBY A-CAL=0.000 LPM Make sure that the T700 is in standby mode. <TST TST> GEN STBY SEQ SETUP X.X...
Calibration and Verification Teledyne API – Model T700 Dynamic Dilution Calibrator 7.4.2.4. Turning O Generator Calibration Points ON / OFF To enable or disable an existing calibration point, press: STANDBY A-CAL=0.000 LPM Make sure that the T700 is in standby mode.
Teledyne API – Model T700 Dynamic Dilution Calibrator Calibration and Verification 7.4.2.5. Performing an Automatic Calibration of the Optional O Generator Note This procedure requires that the T700 calibrator have an optional photometer installed. To run the automatic O generator calibration program, press: STANDBY A-CAL=0.000 LPM...
The procedures described in this section require a separate pressure meter/monitor. The T700 Dynamic Dilution Calibrator has several sensors that monitor the pressure of the gases flowing through the instrument. The data collected by these sensors is used to compensate the final concentration calculations for changes in atmospheric pressure and is stored in the CPU’s memory as various test functions:...
Calibration and Verification Teledyne API – Model T700 Dynamic Dilution Calibrator 7.5.1.1. Calibrating the Diluent, Cal Gas Optional O Generator Pressure Sensors 1. Turn off the calibrator and open the top cover. 2. For the sensor being calibrated, insert a “T” pneumatic connector at the location described in Table 7-2 and shown in Figure 7-6 and Figure 7-7.
1. Turn off the calibrator and open the top cover. 2. Disconnect power to the photometer’s internal pump. 3. Measure the ambient atmospheric pressure of T700’s location in In-Hg-A. 4. Turn on the calibrator and perform the following steps: 5. Turn OFF the T700.
Page 224
Calibration and Verification Teledyne API – Model T700 Dynamic Dilution Calibrator This page intentionally left blank. 06873B DCN6388...
For the most part, the T700 calibrator is maintenance free, there are, however, a minimal number of simple procedures that when performed regularly will ensure that the T700 photometer continues to operate accurately and reliably over its lifetime.
Annually or after Perform Verify Leak any Maintenance 8.2.1 Leak Check Tight or Repair Pneumatic Examine Yes if As needed - - - lines and clean cleaned Only applies to T700 Calibrator’s with O photometer options installed. 06873B DCN6388...
Teledyne API – Model T700 Dynamic Dilution Calibrator Maintenance 8.2. MAINTENANCE PROCEDURES The following procedures are to be performed periodically as part of the standard maintenance of the T700 calibrator. 8.2.1. AUTO LEAK CHECK 8.2.1.1. Equipment Required Four (4) 1/4" Pneumatic caps.
3. Using the 1/8” cap, securely cover the outlet of the internal vent located just behind the valve relay PCA (see Figure 8-1). 4. Use the 1/4" caps to cover the following gas outlet ports on the back of the T700 (see Figure 8-2).
GAS INPUT MANIFOLD OUTLET (on back panel) CAL GAS Cal Gas OUTPUT 1 Mass Flow Controller 2 CAL GAS OUTPUT 2 Volume VENT GAS OUTPUT MANIFOLD Figure 8-4: Gas Flow for Auto-Leak Check Procedure of T700’s with Optional Photometer 06873B DCN6388...
LEAK CHECK PASSED 29.8 PSIG (Zero Air ) bottle EXIT 8.2.1.4. Returning the T700 to Service after Performing an Auto Leak Check 1. Remove all of the caps from the EXHAUST, CAL GAS OUTPUTS (2) and the VENT port and from the internal vent.
Teledyne API – Model T700 Dynamic Dilution Calibrator Maintenance 8.2.2. CLEANING OR REPLACING THE ABSORPTION TUBE Note Although this procedure should never be needed as long as the user is careful to supply the photometer with clean, dry and particulate free zero air only, it is included here for those rare occasions when cleaning or replacing the absorption tube may be required.
Maintenance Teledyne API – Model T700 Dynamic Dilution Calibrator 8.2.3. UV SOURCE LAMP ADJUSTMENT This procedure provides in detail the steps for adjustment of the UV source lamp in the optical bench assembly. This procedure should be done whenever the PHOTO REFERENCE test function value drops below 3000 mV.
Teledyne API – Model T700 Dynamic Dilution Calibrator Maintenance Figure 8-5: Photometer Assembly – Lamp Adjustment / Installation 8.2.4. UV SOURCE LAMP REPLACEMENT This procedure details the steps for replacement of the UV source lamp in the optical bench assembly. This procedure should be done whenever the lamp can no longer be adjusted as described in Section 8.2.3.
Maintenance Teledyne API – Model T700 Dynamic Dilution Calibrator c) Adjust PHOTO_DET within the range of 4400 – 4600 mV. 11. Replace the cover on the calibrator. CAUTION The UV lamp contains mercury (Hg), which is considered hazardous waste. The lamp should be disposed of in accordance with local regulations regarding waste containing mercury.
Teledyne API – Model T700 Dynamic Dilution Calibrator Maintenance Generator Body Adjustment Generator Reference Detector Figure 8-7: Location of O Generator Reference Detector Adjustment Pot 9. Perform the following procedure: 06873B DCN6388...
Page 238
Maintenance Teledyne API – Model T700 Dynamic Dilution Calibrator STANDBY A-CAL=0.000 LPM Make sure that the T700 is <TST TST> GEN STBY SEQ in standby mode. SETUP X.X SEQ CFG CLK PASS MORE EXIT SETUP X.X USER EXIT SETUP X.X...
Page 239
Teledyne API – Model T700 Dynamic Dilution Calibrator Maintenance 11. Replace the calibrator’s cover. 12. Perform an auto-leak check (See Section 8.2.1). 13. Perform an Ozone Generator calibration (see Section 7.4). 06873B DCN6388...
Page 240
Maintenance Teledyne API – Model T700 Dynamic Dilution Calibrator This page intentionally left blank. 06873B DCN6388...
Use common sense when operating inside a running calibrator. 9.1. GENERAL TROUBLESHOOTING The T700 Dynamic Dilution Calibrator has been designed so that problems can be rapidly detected, evaluated and repaired. During operation, it continuously performs diagnostic tests and provides the ability to evaluate its operating parameters without disturbing monitoring operations.
(See Section 9.4.3), the relay PCA (See Section 9.4.7), and the motherboard (See Section9.4.11) be confirmed before addressing the specific warning messages. The T700 will alert the user that a Warning Message is active by flashing the FAULT LED, displaying the Warning message in the Param field along with the CLR button (press to clear Warning message).
Page 243
Teledyne API – Model T700 Dynamic Dilution Calibrator Troubleshooting and Service The calibrator will also alert the user via the Serial I/O COMM port(s) and cause the FAULT LED on the front panel to blink. To view or clear the various warning messages press:...
Page 244
Troubleshooting and Service Teledyne API – Model T700 Dynamic Dilution Calibrator 06873B DCN6388...
One of the calibrator’s improperly vented. MFC PRESSURE mass flow controllers - Leak or blockage exists in the T700’s internal pneumatics WARNING internal gas pressure is <15 PSIG or > 36 PSIG - Failed CAL GAS or DUILUENT pressure sensor...
Regulator pressure is - Incorrectly adjusted O zero air pressure regulator WARNING < 15 PSIG or > 25 PSIG. - Leak or blockage exists in the T700’s internal pneumatics - Failed O Generator Input pressure sensor C Bus failure The CPU cannot...
O concentration. O3GENREF Possible causes of faults are the same as O3 GEN REFERENCE WARNING from Table 9-1. Gas flow problems directly affect the concentration accuracy of the T700’s calibration gas mixtures. O3FLOW - Check for Gas Flow problems.
Troubleshooting and Service Teledyne API – Model T700 Dynamic Dilution Calibrator TEST FUNCTION DIAGNOSTIC RELEVANCE AND CAUSES OF FAULT CONDITIONS. The temperature of the gas in the photometer’s sample chamber is used to calculate the concentration of O in the gas stream. Incorrect sample temperature can cause inaccurate readings.
Teledyne API – Model T700 Dynamic Dilution Calibrator 9.2. USING THE ANALOG OUTPUT TEST CHANNEL The signals available for output over the T700’s analog output channel can also be used as diagnostic tools. See Section 4.7 for instruction on activating the analog output and selecting a function.
Teledyne API – Model T700 Dynamic Dilution Calibrator Troubleshooting and Service 9.3. USING THE INTERNAL ELECTRONIC STATUS LEDS Several LEDs are located inside the instrument to assist in determining if the calibrators CPU, I C bus and Relay PCA are functioning properly.
Teledyne API – Model T700 Dynamic Dilution Calibrator Troubleshooting and Service 9.3.3. VALVE DRIVER PCA STATUS LEDS The Signal I/O submenu also includes VARS that can be used to turn the various input gas valves on and off as part of a diagnostic investigation.
9.4.1. VERIFY SUBSYSTEM CALIBRATION A good first step when troubleshooting the operation of the T700 calibrator is to verify that its major subsystems are properly calibrated. These are: ...
Teledyne API – Model T700 Dynamic Dilution Calibrator Troubleshooting and Service 9.4.3. DC POWER SUPPLY If you have determined that the calibrator’s AC mains power is working, but the unit is still not operating properly, there may be a problem with one of the instrument’s switching power supplies.
Troubleshooting and Service Teledyne API – Model T700 Dynamic Dilution Calibrator A voltmeter should be used to verify that the DC voltages are correct per the values in Table 9-9, and an oscilloscope, in AC mode, with band limiting turned on, can be used to evaluate if the supplies are producing excessive noise (>...
Teledyne API – Model T700 Dynamic Dilution Calibrator Troubleshooting and Service 9.4.6. LCD DISPLAY MODULE Verify the functioning of the front panel display by observing it when power is applied to the instrument. Assuming that there are no wiring problems and that the DC power...
Troubleshooting and Service Teledyne API – Model T700 Dynamic Dilution Calibrator 9.4.9. INPUT GAS PRESSURE / FLOW SENSOR ASSEMBLY The input gas pressure/flow sensor PCA, located at the front of the instrument to the left of the MFCs (see Figure 3-6) can be checked with a Voltmeter.
Teledyne API – Model T700 Dynamic Dilution Calibrator Troubleshooting and Service 9.4.10. PHOTOMETER O3 GENERATOR PRESSURE/FLOW SENSOR ASSEMBLY This assembly is only present in calibrators with O generator and/or photometer options installed. The pressure/flow sensor PCA, located at the rear of the instrument between...
Troubleshooting and Service Teledyne API – Model T700 Dynamic Dilution Calibrator 9.4.11. MOTHERBOARD 9.4.11.1. A/D Functions The simplest method to check the operation of the A-to-D converter on the motherboard is to use the Signal I/O function under the DIAG menu to check the two A/D reference voltages and input signals that can be easily measured with a voltmeter.
Teledyne API – Model T700 Dynamic Dilution Calibrator Troubleshooting and Service For each of the steps the output should be within 1% of the nominal value listed in the table below except for the 0% step, which should be within 0mV ±2 to 3 mV. Ensure you take into account any offset that may have been programmed into channel (See Section 4.10.1.5).
Troubleshooting and Service Teledyne API – Model T700 Dynamic Dilution Calibrator 9.4.11.3. Status Outputs To test the status output electronics: 1. Connect a jumper between the “D“ pin and the “” pin on the status output connector. 2. Connect a 1000 ohm resistor between the “+” pin and the pin for the status output that is being tested.
Teledyne API – Model T700 Dynamic Dilution Calibrator Troubleshooting and Service 9.4.11.4. Control Inputs Table 9-13: T700 Control Input Pin Assignments and Corresponding Signal I/O Functions CONNECTOR INPUT CORRESPONDING I/O SIGNAL CONTROL_IN_1 CONTROL_IN_2 CONTROL_IN_3 CONTROL_IN_4 CONTROL_IN_5 CONTROL_IN_6 Bottom CONTROL_IN_7 Bottom...
Troubleshooting and Service Teledyne API – Model T700 Dynamic Dilution Calibrator Table 9-14: Control Outputs Pin Assignments and Corresponding Signal I/O Functions Check PIN (LEFT TO RIGHT) STATUS CONTROL_OUT_1 CONTROL_OUT_2 CONTROL_OUT_3 CONTROL_OUT_4 CONTROL_OUT_5 CONTROL_OUT_6 CONTROL_OUT_7 CONTROL_OUT_8 CONTROL_OUT_9 CONTROL_OUT_10 CONTROL_OUT_11 CONTROL_OUT_12 9.4.12.
3. Check to ensure the set up command is correct. See Section 6.2. 4. Verify that the Ready to Send (RTS) signal is at logic high. The T700 sets pin 7 (RTS) to greater than 3 volts to enable modem transmission.
Individual control loops are used to maintain the set point of the Photometer UV Lamp (optional), and the Ozone Generator Lamp (optional). If any of these temperatures are out of range or are poorly controlled, the T700 will perform poorly. 9.4.15.1. Box / Chassis Temperature The box temperature sensor is mounted to the Motherboard and cannot be disconnected to check its resistance.
Teledyne API – Model T700 Dynamic Dilution Calibrator Troubleshooting and Service 9.4.15.4. Ozone Generator Temperature There are three possible causes for the Ozone Generator temperature to have failed. The O Gen heater has failed. Check the resistance between pins 5 and 6 on the six-pin connector adjacent to the UV Lamp on the O Generator.
Troubleshooting and Service Teledyne API – Model T700 Dynamic Dilution Calibrator 9.5.1.2. Noisy, Unstable, or Non-Linear Span O Readings Check for leaks in the pneumatic systems as described in Section 8.2.1. Check for proper operation of the meas/ref switching valve as described in Section 9.5.2.
Teledyne API – Model T700 Dynamic Dilution Calibrator Troubleshooting and Service 9.5.2. CHECKING MEASURE / REFERENCE VALVE 1. To check the function of the photometer’s measure / reference valve: 2. Set the calibrator’s front panel display to show the PHOTO REFERENCE test function (see Section 4.1.1).
Troubleshooting and Service Teledyne API – Model T700 Dynamic Dilution Calibrator 9.5.3. CHECKING THE UV LAMP POWER SUPPLY Note A schematic and physical diagram of the Lamp Power Supply can be found in Appendix D. WARNING Hazardous voltage present - use caution.
Teledyne API – Model T700 Dynamic Dilution Calibrator Troubleshooting and Service 9.6. TROUBLESHOOTING THE OPTIONAL O GENERATOR The only significant components of the O generator that might reasonable malfunction is the power supply assembly for the UV source lamp and the lamp itself.
Troubleshooting and Service Teledyne API – Model T700 Dynamic Dilution Calibrator 9.7. SERVICE PROCEDURES 9.7.1. DISK-ON-MODULE REPLACEMENT PROCEDURE Replacing the Disk-on-Module (DOM) will cause loss of all DAS data; it may also cause some of the instrument configuration parameters to be lost unless the replacement DOM carries the exact same firmware version.
9.9. FREQUENTLY ASKED QUESTIONS (FAQs) The following list of FAQs is from the Teledyne API’s Customer Service Department’s most commonly asked questions relating to the T700 Dynamic Dilution Calibrator. Question...
Page 274
Troubleshooting and Service Teledyne API – Model T700 Dynamic Dilution Calibrator This page intentionally left blank. 06873B DCN6388...
(zero air). Using several Mass Flow Controllers (MFCs) the T700 calibrator creates exact ratios of diluent and source gas by controlling the relative rates of flow of the various gases, under conditions where the temperature and pressure of the gases being mixed is known (and therefore the density of the gases).
Page 276
MFCs in real time so the flow rate control can be constantly adjusted to maintain a stable output concentration. The T700 calibrator’s level of control is so precise that bottles of mixed gases can be used as source gas. Once the exact concentrations of all of the gases in the bottle are programmed into the T700, it will create an exact output concentration of any of the gases in the bottle.
Because ozone is a very reactive and therefore under normal ambient conditions a short- lived gas, it cannot be reliably bottled, however, an optional O generator can be included in the T700 calibrator that allows the instrument to be use to create calibration mixtures that include O This ability to generate O...
3 flow = required diluent gas flow flow Again, this is a dynamic process. An optional photometer can be added the T700 calibrator that allows the CPU to tracks the chemiluminescent reaction created when the NO and O interact to measure the decrease in NO concentration as NO is produced.
10.2.1.1. Diluent and Source Gas Flow Control Diluent and source gas flow in the T700 calibrator is a directly and dynamically controlled buy using highly accurate Mass Flow Controller. These MFCs include internal sensors that determine the actual flow of gas though each and feedback control circuitry that uses this data to adjust the flow as required.
Teledyne API – Model T700 Dynamic Dilution Calibrator Principles of Operation 10.2.1.3. Critical Flow Orifices The most important component of the flow control assemblies is the critical flow orifice. Critical flow orifices are a remarkably simple way to regulate stable gas flow rates.
10.2.2. INTERNAL GAS PRESSURE SENSORS The T700 includes a single pressure regulator. Depending upon how many and which options are installed in the T700 calibrator, there are between two and four pressure sensors installed as well. In the basic unit a printed circuit, assembly located near the front of the calibrator near the MFCs includes sensors that measure the pressure of the diluent gas and the source gas currently selected to flow into the calibrator.
Teledyne API – Model T700 Dynamic Dilution Calibrator Principles of Operation 10.3. ELECTRONIC OPERATION 10.3.1. OVERVIEW ANALOG RS232 COM2 USB COM Ethernet Male Female port C Bus) Touchscreen Display Absorption tube Figure 10-3: T700 Electronic Block Diagram 06873B DCN6388...
CPU and the calibrator’s other major components. Data are generated by the various sub components of the T700 (e.g. flow data from the MFCs, O concentration from the optional photometer). Analog signals are converted into digital data by a unipolar, analog-to-digital converter, located on the motherboard.
The DOM is a 44-pin IDE flash chip with a storage capacity up to 256 MB. It is used to store the computer’s operating system, the Teledyne API firmware, and most of the operational data. The LEDs on the DOM indicate power and reading/writing to or from the DOM.
Principles of Operation Teledyne API – Model T700 Dynamic Dilution Calibrator 10.3.3. RELAY PCA The relay PCA is one of the central switching and power distribution units of the calibrator. It contains power relays, valve drivers and status LEDs for all heated zones and valves, as well as thermocouple amplifiers, power distribution connectors and the two switching power supplies of the calibrator.
10.3.3.1. Valve Control The relay PCA also hosts two valve driver chips, each of which can drive up four valves. In the T700, the relay PCA controls only those valves associated with the O generator and photometer options. All valves related to source gas and diluent gas flow are controlled by a separate valve driver PCA (see Section 10.3.4).
Principles of Operation 10.3.4. VALVE DRIVER PCA The valves that operate the T700 calibrator’s main source gas and diluent gas inputs are controlled by a PCA that is attached directly to the input valve manifold (see Figure 3-5 or Figure 3-6). Like the relay PCA, the valve driver PCA communicates with T700’s CPU through the motherboard over the I C bus.
10.3.5.4. Analog Outputs The T700 calibrator comes equipped with one analog output. It can be set by the user to output a signal level representing any one of the test parameters (see Table 4-14) and will output an analog VDC signal that rises and falls in relationship with the value of the chosen parameter.
Teledyne API – Model T700 Dynamic Dilution Calibrator Principles of Operation 10.3.5.5. External Digital I/O The external digital I/O performs two functions. The STATUS outputs carry logic-level (5V) signals through an optically isolated 8-pin connector on the rear panel of the calibrator. These outputs convey on/off information about certain calibrator conditions such as SYSTEM OK.
WARNING The T700 calibrator is equipped with a universal power supply that allows it to accept any AC power configuration, within the limits specified in Table 2-2. Should the power circuit breaker trip correct the condition causing this situation before turning the calibrator back on.
Principles of Operation 10.4. FRONT PANEL TOUCHSCREEN/DISPLAY INTERFACE The most commonly used method for communicating with the T700 Dynamic Dilution Calibrator is via the instrument’s front panel LCD touchscreen display from where users can input data and receive information directly.
Teledyne API – Model T700 Dynamic Dilution Calibrator 10.5. SOFTWARE OPERATION The T700 calibrator’s core module is a high performance, X86-based microcomputer running Windows CE. On top of the Windows CE shell, special software developed by Teledyne API interprets user commands from various interfaces, performs procedures and tasks and stores data in the CPU’s memory devices.
Teledyne API – Model T700 Dynamic Dilution Calibrator Principles of Operation 10.6. O GENERATOR OPERATION 10.6.1. PRINCIPLE OF PHOTOLYTIC O GENERATION Ozone is a naturally occurring substance that is sometimes called "activated oxygen". It contains three atoms of oxygen (O ) instead of the usual two found in normal oxygen ) that is essential for life.
Principles of Operation Teledyne API – Model T700 Dynamic Dilution Calibrator 10.6.2. O3 GENERATOR – PNEUMATIC OPERATION Pneumatic flow through the O generator is created by supplying zero air (diluent) to it under pressure. The zero air source must be capable of maintaining a continuous flow...
Teledyne API – Model T700 Dynamic Dilution Calibrator Principles of Operation 10.6.3. O3 GENERATOR – ELECTRONIC OPERATION Electronically the O generator and its subcomponents act as peripheral devices operated by the CPU via the motherboard. Sensors, such as the UV lamp thermistor send analog data to the motherboard, where it is digitized.
Generator Electronic Components Location 10.6.3.1. O Generator Temperature Control In order to operate at peak efficiency the UV lamp of the T700’s O generator is maintained at a constant 48ºC. If the lamp temperature falls below 43ºC or rises above 53ºC a warning is issued by the calibrators CPU.
20 PSIG on this line. If the pressure drops below 15 PSIG or rises above 25 PSIG a warning is issued. 10.7. PHOTOMETER OPERATION The Model T700 calibrator’s optional photometer determines the concentration of Ozone (O ) in a sample gas drawn through it. Sample and calibration gases must be supplied at ambient atmospheric pressure in order to establish a stable gas flow through the absorption tube where the gas’...
Teledyne API – Model T700 Dynamic Dilution Calibrator Principles of Operation Finally, to convert the result into Parts per Billion (PPB), the following change is made: Equation 10-8 Τ inHg × × × α Ρ Κ The T700 photometer: ...
Teledyne API – Model T700 Dynamic Dilution Calibrator Principles of Operation 10.7.1.3. The Absorption Path In the most basic terms, the T700 photometer uses a high energy, mercury vapor lamp to generate a beam of UV light. This beam passes through a window of material...
Principles of Operation Teledyne API – Model T700 Dynamic Dilution Calibrator 10.7.2. PHOTOMETER LAYOUT The photometer is where the absorption of UV light by ozone is measured and converted into a voltage. It consists of several sub-assemblies: A mercury-vapor UV lamp. This lamp is coated in a material that optically screens the UV radiation output to remove the O producing 185nm radiation.
Teledyne API – Model T700 Dynamic Dilution Calibrator Principles of Operation flow control assembly containing a critical flow orifice directly upstream of the pump but down stream from the absorption tube. The critical flow orifice installed in the pump supply line is tuned to create a flow of 800 /min.
Teledyne API – Model T700 Dynamic Dilution Calibrator 10.7.4.1. O Photometer Temperature Control In order to operate at peak efficiency the UV lamp of the T700’s O photometer is maintained at a constant 58ºC. This is intentionally set at a temperature higher than the ambient temperature of the T700’s operating environment to ensure that local changes in...
A PRIMER ON ELECTRO-STATIC DISCHARGE Teledyne API considers the prevention of damage caused by the discharge of static electricity to be extremely important part of making sure that your analyzer continues to provide reliable service for a long time. This section describes how static electricity occurs, why it is so dangerous to electronic components and assemblies as well as how to prevent that damage from occurring.
A Primer on Electro-Static Discharge Teledyne API – Model T700 Dynamic Dilution Calibrator workbench, using a plastic handled screwdriver or even the constant jostling of Styrofoam pellets during shipment can also build hefty static charges Table 11-1: Static Generation Voltages for Typical Activities...
Teledyne API – Model T700 Dynamic Dilution Calibrator A Primer on Electro-Static Discharge Any time a charged surface (including the human body) discharges to a device. Even simple contact of a finger to the leads of a sensitive device or assembly can allow enough discharge to cause damage.
A Primer on Electro-Static Discharge Teledyne API – Model T700 Dynamic Dilution Calibrator building static electric charges themselves. This does not prevent discharges from static fields built up on other things, like you and your clothing, from discharging through the instrument and damaging it.
Page 311
Teledyne API – Model T700 Dynamic Dilution Calibrator A Primer on Electro-Static Discharge Use metallic anti-ESD bags for storing and shipping ESD sensitive components and assemblies rather than pink-poly bags. The famous, pink-poly bags are made of a plastic that is impregnated with a liquid (similar to liquid laundry detergent) which very slowly sweats onto the surface of the plastic creating a slightly conductive layer over the surface of the bag.
A Primer on Electro-Static Discharge Teledyne API – Model T700 Dynamic Dilution Calibrator 11.4.2. BASIC ANTI-ESD PROCEDURES FOR ANALYZER REPAIR AND MAINTENANCE 11.4.2.1. Working at the Instrument Rack When working on the analyzer while it is in the instrument rack and plugged into a properly grounded power supply 1.
A Primer on Electro-Static Discharge 11.4.2.3. Transferring Components from Rack to Bench and Back When transferring a sensitive device from an installed Teledyne API analyzer to an anti- ESD workbench or back: 1. Follow the instructions listed above for working at the instrument rack and workstation.
A Primer on Electro-Static Discharge Teledyne API – Model T700 Dynamic Dilution Calibrator 11.4.2.5. Packing Components for Return to Teledyne API’s Customer Service Always pack electronic components and assemblies to be sent to Teledyne API’s Customer Service in anti-ESD bins, tubes or bags. WARNING ...
Page 315
10 megabits per second (Mbps) 100BaseT same as 10BaseT except ten times faster (100 Mbps) APICOM name of a remote control program offered by Teledyne-API to its customers ASSY Assembly Code-Activated Switch Corona Discharge, a frequently luminous discharge, at the surface of a...
Page 316
A Primer on Electro-Static Discharge Teledyne API – Model T700 Dynamic Dilution Calibrator Term Description/Definition sulfur dioxide metric abbreviation for cubic centimeter (replaces the obsolete abbreviation “cc”) Central Processing Unit Digital-to-Analog Converter Data Acquisition System Data Communication Equipment Dry Filter Unit Dynamic Host Configuration Protocol.
Page 317
Teledyne API – Model T700 Dynamic Dilution Calibrator A Primer on Electro-Static Discharge Term Description/Definition Internet Protocol Internal Zero Span Local Area Network Liquid Crystal Display Light Emitting Diode Liters Per Minute Mass Flow Controller Measure/Reference the mass, expressed in grams, of 1 mole of a specific substance. Conversely, one mole is the amount of the substance needed for the molar mass to be the same number in grams as the atomic mass of that substance.
Page 318
A Primer on Electro-Static Discharge Teledyne API – Model T700 Dynamic Dilution Calibrator Term Description/Definition ® Teflon Poly Vinyl Chloride, a polymer used for downstream tubing Reading specification and standard describing a serial communication method between RS-232 DTE (Data Terminal Equipment) and DCE (Data Circuit-terminating Equipment)
Page 319
APPENDIX A-1: Models T700 and 700E Software Menu Trees APPENDIX A-2: Models T700 and 700E Setup Variables Available Via Serial I/O APPENDIX A-3: Models T700 and 700E Warnings and Test Measurements Via Serial I/O APPENDIX A-4: Models T700 and 700E Signal I/O Definitions...
Page 320
APPENDIX A – Version Specific Software Documentation (05623D DCN5839) T700, M700E Calibrator Manuals 06873B DCN6388...
Page 325
T700, M700E Calibrator Manuals APPENDIX A-1: Software Menu Trees, Revision D.3 (05623D DCN5839) Figure A-4: PRIMARY SETUP Menu - SOURCE GAS CONFIGURATION Submenu 06873B DCN6388...
Page 326
APPENDIX A-1: Software Menu Trees, Revision D.3 (05623D DCN5839) T700, M700E Calibrator Manuals MAIN MENU SETUP EDIT PRNT EXIT PREV NEXT PRNT EXIT EDIT Inserts a new sequence Edits existing sequence Cycles through list of already programmed calibration sequences <SET SET>...
Page 327
T700, M700E Calibrator Manuals APPENDIX A-1: Software Menu Trees, Revision D.3 (05623D DCN5839) SETUP MORE MAIN MENU COMM DIAG1 FLOW VARS ENTER PASSWORD: 818 COM1 COM2 TARG STAT PREV NEXT JUMP EDIT PRNT SETUP X.X MACHINE ID:700 ID <SET SET>...
Page 329
T700, M700E Calibrator Manuals APPENDIX A-1: Software Menu Trees, Revision D.3 (05623D DCN5839) SETUP MORE DIAG MAIN MENU ENTER PASSWORD: 818 PREV NEXT ANALOG ANALOG I/O AUTO LEAK TEST CHAN PHOTO FLOW O3 GEN PRESSURE SIGNAL I/O Backpressure OUTPUT CONFIGURATION...
Page 331
T700, M700E Calibrator Manuals APPENDIX A-2: Setup Variables For Serial I/O, Revision D.3 (05623D DCN5839) APPENDIX A-2: Setup Variables For Serial I/O, Revision D.3 Table A-1: Setup Variables, Revision D.3 Setup Variable Numeric Default Value Description Units Value Range Low Access Level Setup Variables (818 password) PHOTO_LAMP ºC...
Page 332
APPENDIX A-2: Setup Variables For Serial I/O, Revision D.3 (05623D DCN5839) T700, M700E Calibrator Manuals Setup Variable Numeric Default Value Description Units Value Range PDELTA_GAIN PPB/dIn-Hg -200–200 Multiplied by difference between measure and reference pressure and added to concentration. PDELTA_CAL_DUR Minutes 0.1–20...
Page 333
T700, M700E Calibrator Manuals APPENDIX A-2: Setup Variables For Serial I/O, Revision D.3 (05623D DCN5839) Setup Variable Numeric Default Value Description Units Value Range REF_DELAY Seconds 1–300 generator reference feedback control delay. REF_FREQ Seconds 1–60 generator reference adjustment frequency. REF_FSIZE Samples 1–10...
Page 334
APPENDIX A-2: Setup Variables For Serial I/O, Revision D.3 (05623D DCN5839) T700, M700E Calibrator Manuals Setup Variable Numeric Default Value Description Units Value Range RS232_MODE BitFlag 0–65535 RS-232 COM1 mode flags. Add values to combine flags. 1 = quiet mode...
Page 335
T700, M700E Calibrator Manuals APPENDIX A-2: Setup Variables For Serial I/O, Revision D.3 (05623D DCN5839) Setup Variable Numeric Default Value Description Units Value Range RS232_MODE2 — 0–65535 RS-232 COM2 mode flags. (Same settings as RS232_MODE, plus these when MODBUS option is installed:)
Page 336
APPENDIX A-2: Setup Variables For Serial I/O, Revision D.3 (05623D DCN5839) T700, M700E Calibrator Manuals Setup Variable Numeric Default Value Description Units Value Range TEST_CHAN_ID NONE, Diagnostic analog output ID. Enclose value in double quotes O3 PHOTO (") when setting from the RS-232 MEAS, interface.
Page 337
T700, M700E Calibrator Manuals APPENDIX A-2: Setup Variables For Serial I/O, Revision D.3 (05623D DCN5839) Setup Variable Numeric Default Value Description Units Value Range BOX_SET ºC 0–100 Internal box temperature set point and warning limits. Warnings: 5–45 Molar mass of sample gas for...
Page 338
APPENDIX A-2: Setup Variables For Serial I/O, Revision D.3 (05623D DCN5839) T700, M700E Calibrator Manuals Setup Variable Numeric Default Value Description Units Value Range FACTORY_OPT — 0–65535 Factory option flags. Add values to combine options. 1 = permeation tube #1 installed...
Page 339
T700, M700E Calibrator Manuals APPENDIX A-3: Warnings and Test Functions, Revision D.3 (05623D DCN5839) APPENDIX A-3: Warnings and Test Functions, Revision D.3 Table A-2: Warning Messages, Revision D.3 Name Message Text Description Warnings WSYSRES SYSTEM RESET Instrument was power-cycled or the CPU was reset.
Page 340
APPENDIX A-3: Warnings and Test Functions, Revision D.3 (05623D DCN5839) T700, M700E Calibrator Manuals Name Message Text Description WFRONTPANEL FRONT PANEL WARN Firmware is unable to communicate with the front panel. WMFCCOMM MFC COMMUNICATION WARNING Firmware is unable to communicate with any MFC.
Page 341
T700, M700E Calibrator Manuals APPENDIX A-3: Warnings and Test Functions, Revision D.3 (05623D DCN5839) Table A-3: Test Functions, Revision D.3 Name Message Text Description Test Measurements Parmeter name in T-Series and in E-Series w/ software v. D.3 Parameter name in earlier E-Series software and higher.
Page 342
APPENDIX A-3: Warnings and Test Functions, Revision D.3 (05623D DCN5839) T700, M700E Calibrator Manuals Name Message Text Description TESTCHAN TEST=2753.9 MV Value output to TEST_OUTPUT analog output, selected with TEST_CHAN_ID variable. CLOCKTIME TIME=14:48:01 Current instrument time of day clock. The name is used to request a message via the RS-232 interface, as in “T BOXTEMP”.
Page 343
T700, M700E Calibrator Manuals APPENDIX A-3: Warnings and Test Functions, Revision D.3 (05623D DCN5839) APPENDIX A-4: Signal I/O Definitions Table A-4: Signal I/O Definitions 3 Signal Name Bit or Channel Description Number U11, J1004, control inputs, pins 1-6 = bits 0-5, read, default I/O address 321 hex CONTROL_IN_1 –...
Page 344
APPENDIX A-3: Warnings and Test Functions, Revision D.3 (05623D DCN5839) T700, M700E Calibrator Manuals Signal Name Bit or Channel Description Number Relay board digital output (PCF8575), write, default I C address 44 hex RELAY_WATCHDOG Alternate between 0 and 1 at least every 5 seconds to keep...
Page 345
T700, M700E Calibrator Manuals APPENDIX A-3: Warnings and Test Functions, Revision D.3 (05623D DCN5839) Signal Name Bit or Channel Description Number PURGE_VALVE 1 = open purge valve 0 = close INPUT_VALVE 1 = open input (zero-air) shut-off valve 0 = close...
Page 346
APPENDIX A-3: Warnings and Test Functions, Revision D.3 (05623D DCN5839) T700, M700E Calibrator Manuals Signal Name Bit or Channel Description Number 6–7 Spare Rear board DAC MUX analog inputs DAC_CHAN_1 DAC channel 0 loopback DAC_CHAN_2 DAC channel 1 loopback DAC_CHAN_3...
Page 347
T700, M700E Calibrator Manuals APPENDIX A-5: MODBUS Register Map (05623D DCN5839) APPENDIX A-5: MODBUS Register Map MODBUS Description Units Register Address (dec., 0-based) MODBUS Floating Point Input Registers (32-bit IEEE 754 format; read in high-word, low-word order; read-only) Actual cal. gas flow rate...
Page 348
APPENDIX A-5: MODBUS Register Map (05623D DCN5839) T700, M700E Calibrator Manuals MODBUS Description Units Register Address (dec., 0-based) MODBUS Discrete Input Registers (single-bit; read-only) System reset warning Box temperature warning Photometer lamp temperature warning generator lamp temperature warning Permeation tube #1 temperature warning...
Page 351
APPENDIX B - Spare Parts Use of replacement parts other than those supplied by T-API may result in non Note compliance with European standard EN 61010-1. Due to the dynamic nature of part numbers, please refer to the Website or call Note Customer Service for more recent updates to part numbers.
Page 352
This page intentionally left blank. 06873B DCN6388...
Page 353
T700 Spare Parts List PN 06852B DCN6014 03/10/2011 1 of 3 page(s) Part Number Description 000940100 ORIFICE, 3 MIL, 03 GEN 003290000 THERMISTOR, BASIC (VENDOR ASSY)(KB) 006120100 ASSY, UV LAMP, OZONE GENERATOR 014540300 CONTROLLER, MFC, HFC-212, 100SCCM * 014550300 CONTROLLER, MFC, HFC-212, 10 SLM *...
Page 354
PCA, AUX-I/O BD, ETHERNET, ANALOG & USB 067300100 PCA, AUX-I/O BOARD, ETHERNET 067300200 PCA, AUX-I/O BOARD, ETHERNET & USB 067900000 LCD MODULE, W/TOUCHSCREEN(KB) 068290100 DOM, w/SOFTWARE, STD, T700 * 068730000 MANUAL, T700, OPERATORS 068810000 PCA, LVDS TRANSMITTER BOARD 069500000 PCA, SERIAL & VIDEO INTERFACE BOARD 072150000 ASSY.
Page 355
T700 Spare Parts List PN 06852B DCN6014 03/10/2011 3 of 3 page(s) Part Number Description HW0000356 PAD, THERMAL, TO-220, W/ ADHV HW0000453 SUPPORT, CIRCUIT BD, 3/16" ICOP KIT000253 ASSY & TEST, SPARE PS37 KIT000289 AKIT, UV LAMP P/S PCA, 041660100...
Page 356
T700 RECOMMENDED SPARES STOCKING LEVELS (Reference: 07565A DCN6306) UNITS PART NO. DESCRIPTION 6-10 11-20 21-30 006120100 ASSY, OZONE GEN LAMP 014540300 MASS FLOW CONTROLLER, 100CCM 014550300 MASS FLOW CONTROLLER, 10LPM 040010000 ASS, FAN, REAR PANEL 040030500 PCA, ;PRESS SENSORS (2X), 700E...
Page 357
PHOTO OFFSET PH OFFST 0 ±10 PPB If ozone generator option installed. If photometer option installed. If permeation tube installed. Standard flow TELEDYNE INSTRUMENTS CUSTOMER SERVICE EMAIL: api-customerservice@teledyne.com PHONE: (858) 657-9800 TOLL FREE: (800) 324-5190 FAX: (858) 657-9816 06873B DCN6388...
Page 358
What tests have you done trying to solve the problem? ______________________________________________ __________________________________________________________________________________________ __________________________________________________________________________________________ __________________________________________________________________________________________ __________________________________________________________________________________________ Thank you for providing this information. Your assistance enables Teledyne Instruments to respond faster to the problem that you are encountering. OTHER NOTES: ____________________________________________________________________________ __________________________________________________________________________________________ __________________________________________________________________________________________...
Page 359
APPENDIX D – Wire List and Electronic Schematics 06873B DCN6388...
Page 360
This page intentionally left blank. 06873B DCN6388...
Page 361
Interconnect List T700 (Reference: 069140100A DCN5870) Revision Description Checked Date Production Release 10/4/2010 5870 CONNECTION FROM CONNECTION TO Cable Signal Assembly Assembly Part # 036490100 CBL, AC POWER AC Line Power Entry CN0000073 Power Switch SW0000025 AC Neutral Power Entry...
Page 362
Interconnect List T700 (Reference: 069140100A DCN5870) CONNECTION FROM CONNECTION TO Cable Signal Assembly Assembly Part # 051790000 CBL, POWER & SIGNAL DISTRIBUTION Motherboard 058021400 J109 Bench Det 041200000 AGND Motherboard 058021400 J109 12 Bench Det 041200000 +15V Relay Board 045230100...
Page 363
Interconnect List T700 (Reference: 069140100A DCN5870) CONNECTION FROM CONNECTION TO Cable Signal Assembly Assembly Part # 056310000 CBL, I2C SIGNAL +15V Relay Board 045230100 IZS Lamp Supply 041660100 AGND Relay Board 045230100 IZS Lamp Supply 041660100 Bench Lamp Supply 041660500...
Page 364
Interconnect List T700 (Reference: 069140100A DCN5870) CONNECTION FROM CONNECTION TO Cable Signal Assembly Assembly Part # 06746 CBL, MOTHERBOAD TO CPU RXD(0) CPU PCA 067240000 COM1 Motherboard 058021400 RTS(0) CPU PCA 067240000 COM1 Motherboard 058021400 TXD(0) CPU PCA 067240000 COM1...
Page 367
30R, 50W HEADER 6 THERMISTOR Date Change Description 8/1/02 Initial release for PCA schematic SCH, DC HEATER/THERMISTOR The information herein is the APPROVALS DATE property of API and is submitted in strictest con- DRAWN fidence for reference only. Unauthorized use by anyone for any other purposes is CHECKED SIZE...
Page 369
+15V 1.1K ASCX PRESSURE SENSOR 1.0UF LM4040CIZ S1/S4_OUT S2_OUT S3_OUT 10V_REF +15V ASCX PRESSURE SENSOR MINIFIT6 +15V FLOW SENSOR FM_4 CN_647 X 3 1.0UF +15V LM4040CIZ CON4 SCH, PCA 04003, PRESS/FLOW, 'E' SERIES The information herein is the APPROVALS DATE property of API and is submitted in strictest con- DRAWN...