3.3.2.6. Other Pneumatic Connections

60
SPAN GAS/ZERO AIR CALIBRATION MANIFOLD EXHAUST/VENT LINES
The span and zero air manifolds‟ excess gas should be vented to a suitable vent outside
of the room. This vent should be of large enough internal diameter to avoid any
appreciable pressure drop, and it must be located sufficiently downstream of the output
ports to assure that no ambient air enters the manifold due to eddy currents or back
diffusion.
brn
DILUENT
PRESSURE
SENSOR
CAL GAS
PRESSURE
SENSOR
orn
orn
yel
Figure 3-25: T700 Pneumatic Diagram – Base Unit
The standard T700 Dynamic Dilution Calibrator is equipped with one calibration gas
mass flow controller (flow rate 0 – 100 cm
controller (flow rate 0-10 LPM). See Table 1-1 for the various flow rate options.
Some of the T700 Dynamic Dilution Calibrator‟s optional equipment requires additional
pneumatic connections.
O
GENERATOR OPTION
3
Because ozone (O
) quickly breaks down into molecular oxygen (O
3
gas cannot be supplied in precisely calibrated bottles like other gases such as SO
CO
NO, H
S, etc. The optional O
2
2
Dynamic Dilution Calibrator dynamically generate calibration gas mixtures containing
O
.
3
Additionally a glass mixture volume, designed to meet US EPA guidelines for Gas
Phase Titration (GPT), is included with this option. This chamber, in combination with
Teledyne API – Model T700 Dynamic Dilution Calibrator
brn
Diluent
Mass Flow Controller
yel
yel
yel
Cal Gas
Mass Flow Controller
generator extends the capabilities of the T700
3
On Back Panel
blu
blu
GAS OUTPUT MANIFOLD
3
/min) and one diluent gas mass flow
), this calibration
2
PHOTOMETER
INLET
PHOTOMETER
ZERO IN
PHOTOMETER
ZERO OUT
EXHAUST
PHOTOMETER
OUTLET
CAL GAS
OUTPUT 1
CAL GAS
OUTPUT 2
VENT
, CO,
2
06873E DCN7106