Teledyne API – Model T700 Dynamic Dilution Calibrator
7.3.3. SETUP FOR CALIBRATION OF THE O
Note
This procedure requires external sources for zero air and O
reference photometer.

7.3.3.1. Setup Using Direct Connections

Calibration
Source Gas
connections are
as shown in
Section 3.2
Figure 7-3: External Photometer Validation Setup – Direct Connections
Note
A Minimum of 1.1 LPM is required for the external zero air source.
06873E DCN7106
Calibrating the T700 calibrator‟s optional internal photometer requires a different set up
than that used during the normal operation of the calibrator. There are two ways to
make the connections between these instruments and the T700 calibrator: either with
direct connections or calibration manifolds
Figure 7-3 shows the external zero air and O
photometer connected directly to the fixtures on the back of the T700 Calibrator.
External Zero Air Source
REFERENCE
PHOTOMETER
External O
3
PHOTOMETER INLET
PHOTOMETER OUTLET
PHOTOMETER ZERO IN
PHOTOMETER ZERO OUT
CYL1
CYL2
CYL3
CYL4
DILUENT IN
PHOTOMETER
3
Source
Capped
Capped
EXHAUST
Max Length=3 meters ( or 10 feet)
VENT
CAL OUT
CAL OUT
Calibrator
CAL OUT gas outlet
manifolds/instruments
Calibration and Verification
as an external
3
sources as well as the reference
3
Vent line
If the gas flow on this
line is ≥ 5 SLPM
the minimum O.D.
must be 3/8 inch
connections to
201