Advertisement

Overview of Operating Modes and Basic Operation
Note

Table 4-16: Voltage Tolerances for the TEST CHANNEL Calibration

FULL
SCALE
0.1 VDC
1 VDC
5 VDC
10 VDC
158
MANUAL CALIBRATION OF THE TEST CHANNEL CONFIGURED FOR VOLTAGE
RANGES
For highest accuracy, the voltages of the analog outputs can be calibrated manually.
The menu for manually adjusting the analog output signal level will only appear
if the AUTO-CAL feature is turned off for the channel being adjusted (see Section
4.10.1.6).
Calibration is performed with a voltmeter connected across the output terminals and by
changing the actual output signal level using the front panel buttons in 100, 10 or 1
count increments.
V OUT +
V OUT -
ANALYZER

Figure 4-11: Setup for Calibrating the TEST CHANNEL

ZERO
SPAN VOLTAGE
TOLERANCE
±0.0005V
±0.001V
±0.002V
±0.004V
Teledyne API – Model T700 Dynamic Dilution Calibrator
V IN +
V IN -
Recording
SPAN
TOLERANCE
90 mV
±0.001V
900 mV
±0.001V
4500 mV
±0.003V
4500 mV
±0.006V
Device
MINIMUM
ADJUSTMENT
(1 count)
0.02 mV
0.24 mV
1.22 mV
2.44 mV
06873E DCN7106