Setup For Verification And Calibration The O 7.4.1.1. Setup Using Direct Connections; Generator; Figure 7-5: O 3 Generator Calibration Setup - Direct Connections - Teledyne T700 Operation Manual

Dynamic dilution calibrator
Hide thumbs Also See for T700:
Table of Contents

Advertisement

Calibration and Verification
7.4. CALIBRATING THE O
7.4.1. SETUP FOR VERIFICATION AND CALIBRATION THE O

GENERATOR

Note
7.4.1.1. Setup Using Direct Connections
Calibration
Source Gas
connections
are as shown
in Section 3.2
Figure 7-5: O
208
3
An external reference photometer is required for the procedure.
Figure 7-5 shows the reference photometer connected directly to the fixtures on the back
of the T700 Calibrator.
REFERENCE
PHOTOMETER
PHOTOMETER INLET
PHOTOMETER OUTLET
PHOTOMETER ZERO IN
PHOTOMETER ZERO OUT
CYL1
CYL2
CYL3
CYL4
DILUENT IN
Calibrator
Generator Calibration Setup – Direct Connections
3
Teledyne API – Model T700 Dynamic Dilution Calibrator
GENERATOR
Capped
Capped
Capped
EXHAUST
VENT
CAL OUT
CAL OUT
3
Max Length=3 meters
( or 10 feet)
If the gas flow on
this line is
≥ 5 SLPM the
minimum O.D.
must be 3/8 inch
CAL OUT gas outlet
connections to
manifolds/instruments
06873E DCN7106