Maintenance

8.2.5. OZONE GENERATOR UV LAMP ADJUSTMENT OR REPLACEMENT

.
228
This procedure details the steps for replacement and initial adjustment of the ozone
generator lamp. If you are adjusting an existing lamp, skip to Step 8.
1. Turn off the calibrator.
2. Remove the cover from the calibrator.
3. Locate the O
generator (see Figure 3-6).
3
UV Lamp
Set Screws
O
Generator
3
Body
Figure 8-6: O
3
4. Remove the two setscrews on the top of the O
lamp.
5. Inspect the o-ring beneath the nut and replace if damaged.
6. Install the new lamp in O
7. Tighten the two setscrew
8. Turn on calibrator and allow it to stabilize for at least 30 minutes.
9. Replace the calibrator‟s cover.
10. Perform an auto-leak check (See Section 8.2.1).
11. Perform an Ozone Generator calibration (see Section 7.4).
Teledyne API – Model T700 Dynamic Dilution Calibrator