Calibrating The Diluent, Cal Gas Optional O Generator Pressure Sensors - Teledyne T700 Operation Manual

Dynamic dilution calibrator
Hide thumbs Also See for T700:
Table of Contents

Advertisement

Teledyne API – Model T700 Dynamic Dilution Calibrator
7.5.1.1. Calibrating the Diluent, Cal Gas Optional O
Make sure that the T700 is
Toggle these buttons change the
T700's pressure display to match
installed in-line with the pressure
06873E DCN7106
1. Turn off the calibrator and open the top cover.
2. For the sensor being calibrated, insert a "T" pneumatic connector at the location
described in Table 7-2 and shown in Figure 7-6 and Figure 7-7.
3. Turn on the calibrator and perform the following steps:
STANDBY
in standby mode.
SETUP X.X
SETUP X.X
COMM FLOW
SETUP X.X
Toggle these buttons to
enter the correct
PASSWORD (929)
DIAG
DIAG
DIAG PCAL
DIAG PCAL
the value measured by a
pressure monitor
sensor being calibrated
DIAG PCAL
(see Table 8-1 and
Figures 8-6 & 8-7).
DIAG PCAL
4. Turn OFF the T700.
5. Remove the pressure monitor.
6. Restore the pneumatic lines to their proper connections.
7. Close the calibrator‟s cover.
Generator Pressure Sensors
3
A-CAL=0.000 LPM
<TST
TST>
GEN STBY SEQ
PRIMARY SETUP MENU
GAS
SEQ
CFG
CLK PASS
SECONDARY SETUP MENU
VARS
DIAG
ENTER PASSWORD
8
1
8
SIGNAL I/O
PREV
NEXT
Continue pressing NEXT until ...
PRESSURE CALIBRATION
PREV NEXT
ENTR
DILUENT PRESS: 20.56 PSIG
0
2
0
.5
6
CAL GAS PRESS: 20.56 PSIG
0
2
0
.5
6
O3/PERM PRESS: 20.56 PSIG
0
2
0
.5
6
SAMPLE PRESS: 30.26 IN-HG-A
0
3
0
.2
6
Calibration and Verification
SETUP
MORE
EXIT
EXIT
ENTR
EXIT
ENTR
EXIT
EXIT
ENTR
EXIT
ENTR
EXIT
ENTR
EXIT
EXIT
217