T700 Gas Pressure Sensor Calibration; Table 7-2: T700 Pressure Sensor Calibration Setup - Teledyne T700 Operation Manual

Dynamic dilution calibrator
Hide thumbs Also See for T700:
Table of Contents

Advertisement

Teledyne API – Model T700 Dynamic Dilution Calibrator

7.5. T700 GAS PRESSURE SENSOR CALIBRATION

Note
The procedures described in this section require a separate pressure
meter/monitor.
SENSOR
Diluent Pressure Sensor
Cal Gas Pressure Sensor
O
Regulator Pressure
3
Sensor
(Optional O
Generator)
3
Sample Gas Pressure
Sensor
(Optional O
Photometer)
3
06873E DCN7106
The T700 Dynamic Dilution Calibrator has several sensors that monitor the pressure of
the gases flowing through the instrument. The data collected by these sensors is used to
compensate the final concentration calculations for changes in atmospheric pressure and
is stored in the CPU‟s memory as various test functions:

Table 7-2: T700 Pressure Sensor Calibration Setup

ASSOCIATED
TEST FUNCTION
DIL PRESSURE
CAL PRESSURE
REG PRESSURE
PHOTO SPRESS
PRESSURE MONITOR
UNITS
MEASUREMENT POINT
Insert monitor just before the inlet port of the
PSIG
diluent MFC
Insert monitor just before the inlet port of the
PSIG
cal gas MFC
Insert monitor in line between the regulator
and the O
gas pressure sensor located on
3
PSIG
the O
generator / photometer pressure /
3
flow sensor PCA
Use monitor to measure ambient
atmospheric pressure at the calibrator‟s
IN-HG-A
location.
Calibration and Verification
215