Figure 3-22: Set Up For T700 - Connecting The T700 To A Sample Manifold - Teledyne T700 Operation Manual

Dynamic dilution calibrator
Hide thumbs Also See for T700:
Table of Contents

Advertisement

Teledyne API – Model T700 Dynamic Dilution Calibrator
ROOF
TEE
H
O TRAP
2
T700 output flow rate
must be greater than the
Total Flow requirements
of the entire system,
whichever is higher.
VENT & extra
CAL OUT
gas outlets
MUST be capped
Figure 3-22: Set up for T700 – Connecting the T700 to a Sample Manifold
• This is the recommended method for connecting the T700 calibrator to a
Note
• The manifolds as shown in the above drawing are oriented to simplify the
• When initiating calibration, wait a minimum of 15 minutes for the calibrator to
06873E DCN7106
CONNECTING THE CALIBRATOR TO A SAMPLE GAS MANIFOLD
Use this setup when connecting the T700 calibrator to an analyzer network using a
sample manifold. In this case, the sampling cane and the manifold itself act as the vent
for the T700.
Minimum O.D. of this gas line
must be 3/8 inch
CAL OUT
CYL1
CYL2
CAL OUT
CYL3
VENT
CYL4
T700
Mass Flow
DILUENT IN
Calibrator
Teledyne API MODEL
701
Zero Gas Generator
system with analyzers that DO NOT have internal zero/span valves.
drawing.
Their actual orientation in your set-up is with the ports facing
upward. All unused ports must be capped.
flood the entire sampling system with calibration gas.
SAMPLING CANE
FILTER SCREEN
BLOWER / PUMP
SAMPLE
MANIFOLD
Getting Started
VENT TO OUTSIDE OF ROOM
Sample Inlet
Exhaust
Teledyne API
Gas Analyzer
Sample Inlet
Exhaust
Teledyne API
Gas Analyzer
CAL
SOURCE
GAS
57