Overview of Operating Modes and Basic Operation

4.11.3. LEVELS

4.11.4. ACTIVATING AN EXISTING LEVEL

Make sure the T700 is
in standby mode
The T700 will continue
generating calibration gas
at the LEVEL chosen until
the STBY button is
pressed.
162
A LEVEL is a combination of several parameters:
An ID number for the LEVEL
An action, (e.g. GENERATE, GPT, GPTPS & MANUAL)
A target concentration value
An output flow rate (if applicable)
Configuration for one or both of two status output blocks.
Up to twenty levels can be defined and used with the T700 using a range of ID numbers
from 0-98. Level 99 is reserved for standby. The levels are not time based and do not
include characteristics such as start time or duration, therefore a single LEVEL can not
switch between different concentration levels and flow rates.
concentration outputs must be programmed into separate LEVELs which are then
individually started and stopped either by an operator at the calibrator‟s front panel or
through a serial data operation over the RS-232 or Ethernet ports.
To activate an existing defined LEVEL, press:
STANDBY
<TST
STANDBY
STANDBY
MANUAL
<TST
Teledyne API – Model T700 Dynamic Dilution Calibrator
A-CAL=0.0000 LPM
TST>
STBY SEQ
GEN
SYSTEM RESET
AUTO MAN PURG GPT GPTS
LVL
LEVEL:0
0
0
Toggle these buttons until the designation of the
existing defined level program is reached.
ACT CAL=3.000 LPM
TST> GEN STBY
Separate flow and
SETUP
EXIT
ENTR
EXIT
ACT>
SETUP
06873E DCN7106