Power Supply And Circuit Breaker; Figure 10-9: T700 Power Distribution Block Diagram - Teledyne T700 Operation Manual

Dynamic dilution calibrator
Hide thumbs Also See for T700:
Table of Contents

Advertisement

Teledyne API – Model T700 Dynamic Dilution Calibrator

10.3.7. POWER SUPPLY AND CIRCUIT BREAKER

The T700 calibrator is equipped with a universal power supply that allows it to accept
Should the power circuit breaker trip correct the condition causing this situation before
SENSOR SUITES
ANALOG SENSOR
Photometer UV Detector
MFC3
2nd Cal Gas
(Optional)
MFC2
Cal Gas
MFC1
(Diluent)
GAS
TEMPERATURE
SENSORS
GAS
PRESSURE
SENSORS
Photometer
UV Lamp P/S
O
Generator UV
3
Lamp Xformer
O
Generator UV
3
Lamp P/S
06873E DCN7106
The T700 calibrator operates in two main AC power ranges: 100-120 VAC and 220-240
VAC (both ± 10%) between 47 and 63 Hz. A 5-ampere circuit breaker is built into the
ON/OFF switch. In case of a wiring fault or incorrect supply power, the circuit breaker
will automatically turn off the calibrator.
any AC power configuration, within the limits specified in Table 2-2.
Sensor Control
& I/O Logic
Pre-Amplifiers
& Amplifiers
RELAY
PCA
Solenoid
Drivers
Cooling
OPTIONAL VALVES
GPT valve,
O
Gen valve
3
Photometer M/R
valve, etc.)
O
Generator
3
UV Lamp

Figure 10-9: T700 Power Distribution Block diagram

WARNING
turning the calibrator back on.
LOGIC DEVICES
2
(e.g. CPU, I
C bus,
Motherboard, etc.)
+5 VDC
±15 VDC
(+12 VDC)
Fan
Photometer
Pump
VALVE
DILUENT
VALVE
DRIVE PCA
PURGE
VALVE
Principles of Operation
KEY
AC POWER
DC POWER
POWER IN
PS 1
PS 2
CAL GAS 1
VALVE
CAL GAS 2
Solenoid
VALVE
Drivers
CAL GAS 3
VALVE
CAL GAS 4
VALVE
AC
ON / OFF
SWITCH
279

Advertisement

Table of Contents
loading

Table of Contents