Figure 3-24: Set up for T700 – Connecting the T700 to a Dual Span Gas / Zero Air Manifold
This set up is subject to the same notes and conditions as the single calibration
Note
manifold described previously with the following two exceptions:
• The T700 total gas flow rate (Cal Gas Flow Rate + Diluent Flow Rate) out should
be greater than the Total Flow requirements of the entire system.
• The manifolds as shown in the above drawing are oriented to simplify the
drawing. Their actual orientation in your set-up is with the ports facing upward.
All unused ports must be capped.
06873E DCN7106
CALIBRATION MANIFOLD EXHAUST/VENT LINE
The manifold‟s excess gas should be vented outside of the room. This vent should be of
large enough internal diameter to avoid any appreciable pressure drop, and it must be
located sufficiently downstream of the output ports to assure that no ambient air enters
the manifold due to eddy currents or back diffusion.
CONNECTING THE CALIBRATOR TO A DUAL SPAN GAS / ZERO AIR
CALIBRATION MANIFOLD
Another type of calibration setup utilizes separate span gas and the zero air manifolds
(see Figure 3-24).
ROOF
TEE
H
O TRAP
2
VENT & extra CAL OUT gas
outlets MUST be capped
VENT
CYL1
CYL2
CAL OUT
CYL3
CAL OUT
CYL4
T700
Mass Flow
DILUENT IN
Calibrator
5 SLPM
SAMPLING CANE
FILTER SCREEN
BLOWER / PUMP
SAMPLE
MANIFOLD
1
SPAN CAL MANIFOLD
ZERO CAL MANIFOLD
If Gas flow on these lines is
≥ 5 SLPM
1
the minimum O.D. must be 3/8 inch
Getting Started
VENT TO OUTSIDE OF ROOM
Sample Inlet
Exhaust
Span Gas Inlet
Zero Air Inlet
Teledyne API
Gas Analyzer
Sample Inlet
Exhaust
Span Gas Inlet
Zero Air Inlet
Teledyne API
Gas Analyzer
1
59