Teledyne API – Model T700 Dynamic Dilution Calibrator
Principles of Operation

Figure 10-15: O

Generator Electronic Components Location
3
10.6.3.1. O

Generator Temperature Control

3
In order to operate at peak efficiency the UV lamp of the T700‟s O
generator is
3
maintained at a constant 48ºC. If the lamp temperature falls below 43ºC or rises above
53ºC a warning is issued by the calibrators CPU.
This temperature is controlled as described in the section on the relay PCA (Section
10.3.3). The location of the thermistor and heater associated with the O
generator is
3
shown in Figure 10-16: