For highest accuracy, the voltages of the analog outputs can be manually calibrated with a
voltmeter connected across the output terminals (see Figure 3-30) and by changing the
actual output signal level using the Manual Adjust buttons in 100, 10 or 1 count increments
(Figure 3-31).

Figure 3-30. Setup for Manually Calibrating the Analog Output

Table 3-4. Voltage Tolerances for the Analog Output Calibration

FULL
SCALE
TOLERANCE
0.1 VDC
1 VDC
5 VDC
10 VDC

Figure 3-31. Analog Output Manual Calibration Page

083730750B DCN8606
+ D C
G n d
V OUT +
V OUT -
ANALYZER
ZERO
SPAN
VOLTAGE
±0.0005V
90 mV
±0.001V
900 mV
±0.002V
4500 mV
±0.004V
4500 mV
V
Volt
Meter
V IN +
250 Ω
V IN -
Recording
Device
SPAN
ADJUSTMENT
TOLERANCE
±0.001V
±0.001V
±0.003V
±0.006V
MINIMUM
(1 COUNT)
0.02 mV
0.24 mV
1.22 mV
2.44 mV
79