Connecting Gas Outputs from the Calibrator

S
ETUP FOR
Use this setup if you are connecting the T750 calibrator directly to other instruments
without the use of any shared manifolds.
CAL
SOURCE
GAS
Figure 2-12. Set up for T750 – Connecting the Basic T750 to a Sample Manifold
To determine if the gas flow on the vent line is ≥ 5 SLPM, subtract the gas flow for each
instrument connected to the outlets of the T750 from the TOTAL FLOW setting for the
calibrator (see Section 2.5.5).
If the T750 has the optional O
3
800 cc
/min (0.8 SLPM) of additional flow (see Section 2.5.5 or Figure 2-18).
EXAMPLE: Your system has two analyzers each requiring 2 SLPM of cal gas flow and
the T750 includes the O
10 SLPM:
10 SLPM – 2 SLPM – 2 SLPM - 0.8 SLPM = 5.2 SLPM
Therefore, the vent would require a gas line with an O.D. 3/8 inch.
36
CAUTION – ASPHYXIATION HAZARD
Exhaust and Cal Gas Out ports must be vented properly. Carefully
adhere to the pneumatic connection instructions
D
C
IRECT
ONNECTIONS TO
EXHAUST
CYL IN
DILUENT IN
CALGAS OUT
T750
Teledyne API
MODEL 701
Zero Gas Generator
photometer installed, remember that this option requires
3
photometer. If the TOTAL FLOW rate for the calibrator is set at
3
O
I
THER
NSTRUMENTS
Operator must vent T750 CALGAS OUT line to
outside of room, with at least 3 ft of 1/4 inch
tubing. If the gas flow on this line is ≥ 5 SLPM the
minimum O.D. must be 3/8 inch.
Sample Inlet
Teledyne API
Gas Analyzer
Sample Inlet
Teledyne API
Gas Analyzer
083730750B DCN8606