Teledyne API – Model T700 Dynamic Dilution Calibrator
7.4.2. O3 GENERATOR CALIBRATION PROCEDURE
7.4.2.1. Viewing O
3
06873B DCN6388
The T700 calibrator's software includes a routine for automatically calibration the O
generator. A table of drive voltages stored in the T700's memory is the basis for this
calibration. For each point included in the table used by the T700 to calibrate the
optional O
generator the user can set a drive voltage and a dwell time for that point.
3
Each point can also be individually turned OFF or ON.
Generator Calibration Points
To view these calibration points, press:
Calibration and Verification
213