Setup For Verification And Calibration Of The T700'Smfc's; Figure 7-1: Location Of Mfc Outlet Ports - Teledyne T700 Operation Manual

Dynamic dilution calibrator
Hide thumbs Also See for T700:
Table of Contents

Advertisement

Calibration and Verification
7.2.1. SETUP FOR VERIFICATION AND CALIBRATION OF THE T700'S
MFC'S
Note
Input Gas
Pressure
Sensor
PCA
Front Panel
Outlet Port for
Optional 2nd Cal Gas
Mass Flow Controller
200
A separate flow meter is required for the procedure.
1. Turn off the T700 Dynamic Dilution Calibrator.
2. Open the front panel to the T700 calibrator. This is the easiest access to the MFC
output ports.
A locking screw located at the top center of the front panel (See Figure 3-1)
must be removed before the panel can be opened.
3. Attach the flow meter directly to the output port of the MFC to be checked/tested.
Outlet Port for
Cal Gas
Mass Flow Controller

Figure 7-1: Location of MFC Outlet Ports

4. Turn the T700 Dynamic Dilution Calibrator ON.
Teledyne API – Model T700 Dynamic Dilution Calibrator
Outlet Port for
Diluent
Mass Flow Controller
GPT
Chamber
GPT
Valve
ON / OFF
Switch
06873B DCN6388