PART 2: DETERMINE THE AMOUNT OF NO OUTGASSED BY THE NO2 à NO
CONVERTER.
3. Bypass the NO2 à NO converter by placing a short piece of tubing in the gas stream in
place of the converter.
4. Perform a straight dilution with 450 ppb NO gas & air as a diluent gas.
5. Input the NO gas into the analyzer.
6. Allow the machine to stabilize & write down the NOx value on line 2 of GPT data sheet
(Section 4.5.1.1).
7. Remove the converter bypass so that the NO gas is flowing through the NO2 à NO
converter
8. Allow the machine to stabilize.
9. Write down your NOx value on your data sheet on lines 3 AND line 5 of the GPT data sheet.
10. Subtract line 2 from line 3 & write that number down on line 4. Also write the NO value on
line 8 of the data sheet.
The specification shown on the data sheet is the value that is used by Teledyne API
when performing the procedure on new NO
Older NO
specification might be in order.
If this value is a constant or changes only slightly over time, this is not a problem the
machine will calibrate this out.
PART 3: PERFORM THE SIMPLIFIED GPT CALCULATION.
11. Generate the same 450 ppb NO gas & input 400 ppb of O
ppb NO
12. Allow the instrument to stabilize for 10 minutes.
13. Write down the NOx value on line 6 & the NO value on line 9.
14. Subtract line 6 from line 5 & put that onto line 7.
15. Subtract line 8 from line 7 & put that onto line 10.
16. Write the number from line 7 into the blank next to letter A on line 11 & put the number from
line 10 into the blank next to letter B on line 11.
17. Divide A by B & multiply it by 100.
18. Write this value it into the blank next to letter C on lines 11 and 12.
19. Subtract that value from 100 & write it in the blank next to the letter D on line 12.
20. This is the converter efficiency.
This value should be >96%.
083730200A DCN7962
à NO converters might outgas a bit more NO, therefore a slightly wider
2
, if that's what your calibrator says).
2
Teledyne API T200 NO/NO
à NO converters.
2
(or generate 450 ppb NO & 400
3
/NO
Analyzer with NumaView™ Software
2
x
103

Advertisement

Table of Contents
loading

Table of Contents