Thermocouple Inputs And Configuration Jumper (Jp5); Figure 6-15. Heater Control Loop Block Diagram - Teledyne T200 User Manual

No/no2/nox analyzer with numaview software
Hide thumbs Also See for T200:
Table of Contents

Advertisement

Thermistor(s) –
J-type
Thermocouple
( O
Destruct)
3
Not used
on the T200
Note

THERMOCOUPLE INPUTS AND CONFIGURATION JUMPER (JP5)

Although the Relay PCA supports two thermocouple inputs, these analyzers only utilize
one. It is used to sense the temperature of the Ozone Destruct.
This single thermocouple input is plugged into the TC1 input (J15).
TC2 (J16) is currently not used (see Figure 6-13 for location of J15 and J16).
The type and operating parameters of this thermocouple are set using a jumper plug
(JP5).
The default configuration for this thermocouple is:
Type-K
Temperature compensated for Type-K
Isolated
083730200A DCN7962
Reaction Cell, Optional Internal
Span Gas Generator, Optional O
2
Sensor)
RELAY PCA
THERMOCOUPLE
CONFIGURATION
JUMPER
(JP5)
DC Control Logic
(for DC heaters)
Reaction Cell
Heater

Figure 6-15. Heater Control Loop Block Diagram.

The PMT temperature is maintained by a separate control loop
that does not involve the relay PCA (see Section 6.5.2).
Teledyne API T200 NO/NO
MOTHER BOARD
Preamplifiers
and Signal
Conditioning
Cold Junction
Compensation
Solid State
AC Relays
NO
à NO
Other Optional
2
Converter
AC HEATERs
Heater
(Internal Span Gas
Generator;O2 Sensor)
/NO
Analyzer with NumaView™ Software
2
x
Thermistor
interface
A/D
Converter
(V/F)
CPU
205