FUNCTIONAL CHECKS

After warm-up, verify that the software properly supports any hardware options that are
installed (Setup>Instrument menu), and that the instrument is functioning within allowable
operating parameters. Check the Dashboard page against the instrument's Final Test and
Validation Data sheet, which lists these values as they appeared before the instrument left
the factory. (If any functional parameters are not displayed, configure the Dashboard
through the Setup>Dashboard menu to add them; see Section 2.4.2).
These functions are also useful tools for diagnosing problems (information provided in
Section 5.7.2).

CALIBRATION

Before operation begins, the analyzer requires zero and span calibrations, and possibly
HVPS adjustment, followed by calibration of the converter efficiency (CE) points. Also,
any time an analyzer is moved or its configuration changed, it must be calibrated. The
method for performing a calibration differs slightly depending on whether or not any of the
available internal zero air or valve options are installed. Follow the appropriate calibration
instructions presented in Section 4.
083730200A DCN7962

Figure 2-32. Sample Dashboard Page

Teledyne API T200 NO/NO
/NO
Analyzer with NumaView™ Software
2
x
55