Table 4-1. Auto Cal States; Table 4-2. Auto Cal Setup Combinations - Teledyne T200 User Manual

No/no2/nox analyzer with numaview software
Hide thumbs Also See for T200:
Table of Contents

Advertisement

Table 4-1. AUTO CAL States

MODE NAME
Enabled
Calibrate
Zero
High
Table 4-2 shows how the selection boxes would be enabled/disabled for calibration checks
and calibrations.

Table 4-2. Auto Cal Setup Combinations

MODE
Zero
Low
High
Zero Low High
For each sequence, there are four parameters that control operational details: Date, Time
(both in the Start field), Interval, and Duration, as presented in Table 4-3.
100
Teledyne API T200 NO/NO
enables the sequence;
disables the sequence.
enables an actual calibration when the Enabled box is also
allows a calibration check when the Enabled box is also
causes the sequence to perform a Zero calibration when both the
Calibrate and Enabled boxes are also
causes a Zero check when the Enabled box is also
Calibrate box is unchecked (
disables Zero calibration and check
causes the sequence to perform a Low Span calibration when both
the Calibrate and Enabled boxes are also
Low
causes a Low Span check when the Enabled box is also
the Calibrate box is unchecked (
disables Low Span calibration and check
causes the sequence to perform a High Span concentration
calibration when both the Calibrate and Enabled boxes are also
.
causes a High Span check when the Enabled box is also
the Calibrate box is unchecked (
disables the High Span calibration and check.
ACTION
Check
Calibrate
Check
Calibrate
Check
Calibrate
Check
Calibrate
/NO
Analyzer with NumaView™ Software
2
x
ACTION
).
).
).
STATE
Enabled
Calibrate
Zero
.
.
.
and the
.
and
and
Low
High
083730200A DCN7962