Table 6-5. Thermocouple Configuration Jumper (JP5) Pin-Outs

JUMPER
TC INPUT
PAIR
1 – 11
2 – 12
TC1
3 – 13
4 – 14
5 – 15
TC2
NOT USED
ATTENTION

Figure 6-16. Thermocouple Configuration Jumper (JP5) Pin-Outs

206
Teledyne API T200 NO/NO
DESCRIPTION
Gain Selector
Output Scale Selector
Type J Compensation
Type K Compensation
Termination Selector
COULD DAMAGE INSTRUMENT AND VOID WARRANTY
The correct Thermocouple Type must be used if there is ever the
need for replacement. If in doubt please consult Teledyne API
Technical Support.
TC2
Not Used
Purple Jumpers
/NO
Analyzer with NumaView™ Software
2
x
FUNCTION
Selects preamp gain factor for J or K TC
OUT = K TC gain factor;
IN = J TC gain factor
Selects preamp gain factor for J or K TC
OUT = 10 mV / °C; IN = 5 mV / °C
When present, sets Cold Junction Compensation
for J type Thermocouple
When present, sets Cold Junction Compensation
for K type Thermocouple
Selects between Isolated and grounded TC
IN = Isolate TC;
OUT = Grounded TC
TC1
083730200A DCN7962