24. Reconnect all electrical and pneumatic connections.
25. Leak check the system (see Section 5.6.11).
26. Turn ON the analyzer.
27. Verify the basic operation of the analyzer using the ETEST and OTEST features
(Section 5.7.9.11) or zero and span gases, then carry out a hardware calibration
of the analyzer followed by a zero/span point calibration Section 4.2.1).
REMOVING / REPLACING THE RELAY PCA FROM THE INSTRUMENT
This is the most commonly used version of the Relay PCA. It includes a bank of solid state
AC relays. This version is installed in analyzers where components such as AC powered
heaters must be turned ON & OFF.
A retainer plate is installed over the relay to keep them securely seated in their sockets.
The Relay retainer plate installed on the relay PCA covers the lower right mounting screw
of the relay PCA. Therefore, when removing the relay PCA, the retainer plate must be
removed first.
083730200A DCN7962
Figure 5-21. Relay PCA with AC Relay Retainer In Place
Figure 5-22. Relay PCA Mounting Screw Locations
Teledyne API T200 NO/NO
/NO
Analyzer with NumaView™ Software