Table of Contents

Advertisement

4. Remove the integrated screws with a Phillips screw driver and remove the
scrubber manifold from the chassis.
5. Using a 9/16" wrench, remove both fittings from the cartridge.
6. Discard the glass wool.
7. Pour the contents of the scrubber manifold onto a sheet of white paper. If
necessary, remove the plug to ensure that all the contents are poured out.
116
Teledyne API T200 NO/NO

Figure 5-8. Ozone Cleanser Assembly

Notice any discoloration of the contents, which is usually white and slightly
transparent.
The amount of discolored chemical (usually with yellow tint) may give you an
indication of the lifetime of the chemical in your application.
The maintenance cycle of this item is dependent on ambient moisture, sub-
micron particle load and other factors and may differ from that shown in Table
5-1.