The Utilities menu has a variety of functions as described next in Table 2-11.

Table 2-11. Utilities Submenu Descriptions

UTILITIES

MENU
Datalog View
Displays the data logs that were configured via the Setup>Data Logging menu. From this list
a log can be selected and filters applied to view the desired data. (For details on setting up and
running the Data Logger, see Section 2.5.1).
Alerts Log
Displays a history of alerts that are triggered by factory-defined and user-defined Events, such
as warnings and alarms (See Section 2.5.2 for Events configuration).
USB Utilities
Serves multiple purposes using a flash drive connected to the instrument's front panel USB
port:
• download data from the instrument's Data Acquisition System (DAS), the Data Logger, to
a flash drive (Section 2.5.1.3)
• update firmware (Section 5.3)
• transfer instrument configuration from/to other same-model instruments (Section 2.6)
• download a basic operation functionality report (Section 5.3).
Diagnostics
Provides access to various pages that facilitate troubleshooting.
Analog Inputs
Analog Outputs
Digital Inputs
Digital Outputs
Flow Cal
OE Test
O3 Gen Override Used to override the Ozone Generator state when needed, such as for
S
ETUP
The Setup menu is used to configure the instrument's various features, functions, and data
log. Section 2.5 provides details for the menus under Setup.
62
Teledyne API T200 NO/NO
DESCRIPTION
Measure voltage signals of several analog input parameters, including
those from other instrumentation when the External Analog Inputs Option
(Section 2.3.1.2) is installed. These can be logged in the internal data
acquisition system (DAS), by configuring the Data Logger in the
Setup>Data Logging menu (Section 2.5.1).
Show the voltage signals for the functions selected and configured in the
Setup>Analog Outputs menu (Section 2.5.8). Rear panel connections
were presented in Section 2.3.1.3.
Show whether specific available features are turned ON or OFF; for
example, whether or not Maintenance Mode input or Language selection
can be controlled through the front panel, or whether a zero or span
calibration can be activated remotely when an external source is
connected to the rear panel Control In connector (Section 2.3.1.6).
Show the function of user-specified parameters configured in the
Setup>Digital Outputs menu (Section 2.5.7).
Used to calibrate the sample gas flow reading with actual flow measured
by an external device. (See Section 5.6.11.3).
Used to run either an electrical test (ETest) or an optics test (OTest)
(Section 5.7.9.11).
service (Section 5.7.9.15).
/NO
Analyzer with NumaView™ Software
2
x
083730200A DCN7962