NON-LINEAR RESPONSE

The analyzer was factory calibrated to a high level of NO and should be linear to within
1% of full scale. Common causes for non-linearity are:
Leaks in the pneumatic system:
Leaks can add ambient air, zero air or span gas to the current sample gas stream,
which may be changing in concentrations as the linearity test is performed.
Check for leaks as described in Section 5.6.11.
The calibration device is in error:
Check flow rates and concentrations, particularly when using low concentrations.
If a mass flow calibrator is used and the flow is less than 10% of the full scale
flow on either flow controller, you may need to purchase lower concentration
standards.
The standard gases may be mislabeled as to type or concentration.
Labeled concentrations may be outside the certified tolerance.
The sample delivery system may be contaminated.
Check for dirt in the sample lines or reaction cell.
Calibration gas source may be contaminated (NO
Dilution air contains sample or span gas.
Ozone concentration too low because of wet air in the generator.
Generator system needs to be cleaned and dried with dry supply air.
Check the dryer for leaks.
This mostly affects linearity at the low end.
Ozone stream may be contaminated with impurities.
An exhausted ozone cleanser chemical will let compounds such as HNO
ammonia derivatives break through to the reaction cell.
Check the contents of the ozone cleanser and replace as necessary (Section
5.6.3).
This also will affect linearity mostly at the low level.
Sample inlet may be contaminated with NOx exhaust from this or other analyzers.
Verify proper venting of the pump exhaust.
Span gas overflow is not properly vented and creates a back-pressure on the
sample inlet port.
Also, if the span gas is not vented at all and does not supply enough sample gas,
the analyzer may be evacuating the sample line.
Ensure to create and properly vent excess span gas.