6. When complete, as indicated in the Status field, press the Done button, which
7. Power off and restart the instrument to complete the new firmware installation.
I
NSTRUMENT
This section applies to instruments shipped before January 2017. Although unlikely, if ever
the touchscreen appears unresponsive or responds incorrectly, the screen can be calibrated
via the Setup>Instrument>Display Settings menu.
1. Connect a mouse to either of the front panel USB ports.
2. Navigate with the pointer to Setup>Instrument>Display Settings.
3. Click on "Calibrate Touch" and a crosshair appears in the center of the display screen.
4. Note that a timer function is enabled, allowing only 15 seconds to start the calibration
process. If the timer expires, the instrument will exit the calibration screen and return to
normal operation.
5. Click the very center of the crosshair.
6. When a new crosshair appears in the upper left corner of the screen, carefully and
accurately click and hold the very center of that crosshair until it finishes shrinking, then
release.
7. Repeat Step 6 for each of the corners.
8. Once the process is completed, a CANCEL and an ACCEPT button appear in the lower
left corner: Test the accuracy of the calibration by touching parts of the screen and see
that the mouse pointer follows your touches.
9. If you press the CANCEL button, the calibration won't be altered. Otherwise, press the
ACCEPT button.
083730200A DCN7962
replaces the Cancel button, and remove the flash drive.
D
C
ISPLAY
ALIBRATION

Figure 5-4. Touchscreen Calibration Page

Teledyne API T200 NO/NO
(
FOR EARLIER INSTRUMENTS
/NO
Analyzer with NumaView™ Software