Connecting Gas Outputs From The Calibrator; Figure 3-17: Set Up For T700 - Connecting The Basic T700 To A Sample Manifold - Teledyne T700 Operation Manual

Dynamic dilution calibrator
Hide thumbs Also See for T700:
Table of Contents

Advertisement

Teledyne API – Model T700 Dynamic Dilution Calibrator

3.3.2.5. Connecting Gas Outputs from the Calibrator

Figure 3-17: Set up for T700 – Connecting the Basic T700 to a Sample Manifold
06873B DCN6388
SET UP FOR DIRECT CONNECTIONS TO OTHER INSTRUMENTS
Use this setup if you are connecting the T700 calibrator directly to other instruments
without the use of any shared manifolds.
To determine if the gas flow on the vent line is ≥ 5 SLPM subtract the gas flow for each
instrument connected to the outlets of the T700 from the TOTAL FLOW setting for
the calibrator (see Section 3.4.9).
If the T700 has the optional O3 photometer installed remember that this option requires
800 cc3/min (0.8 LPM) of additional flow (see Section 3.4.9 or Figure 3-23).
EXAMPLE: Your system has two analyzers each requiring 2SLPM of cal gas flow and
the T700 includes the O
calibrator is set at 10 SLPM:
10LPM - 2LPM - 2LPM - 0.8 LPM = 5.2LPM
Therefore, the vent would require a gas line with an O.D. 3/8 inch.
photometer. If the TOTAL FLOW rate for the
3
Getting Started
59

Advertisement

Table of Contents
loading

This manual is also suitable for:

T700u

Table of Contents