Overview of Operating Modes and Basic Operation

4.10.1.6. Test Channel Calibration

NOTE:
TEST CHANNELS
configured for 0.1V full
scale should always be
calibrated manually.
Toggle this button
to turn AUTO CAL
ON or OFF
(OFF = manual
calibration mode).
160
TEST CHANNEL calibration needs to be carried out on first startup of the calibrator
(performed in the factory as part of the configuration process) or whenever recalibration
is required. The analog outputs can be calibrated automatically or adjusted manually.
In its default mode, the instrument is configured for automatic calibration of all
channels, which is useful for clearing any analog calibration warnings associated with
channels that will not be used or connected to any input or recording device, e.g., data
logger.
Manual calibration should be used for the 0.1V range or in cases where the outputs must
be closely matched to the characteristics of the recording device. Manual calibration
requires the AUTOCAL feature to be disabled.
ENABLING OR DISABLING THE TEST CHANNEL AUTOCAL FEATURE
To enable or disable the AUTOCAL feature for the TEST CHANNEL, press:
AIO CONFIGURATION SUBMENU
DIAG
PREV NEXT
DIAG AIO
SET>
Continue pressing SET> until you reach the
DIAG AIO
<SET SET>
DIAG AIO
SET>
Continue pressing SET> until ...
DIAG AIO
<SET SET>
DIAG AIO
ON
DIAG AIO
OFF
Teledyne API – Model T700 Dynamic Dilution Calibrator
From the
(See Section 6.9.1.1.)
ANALOG I/O CONFIGURATION
ENTR
AOUTS CALIBRATED: NO
CAL
output to be configured
TEST_OUTPUT: 5V, OVR, NOCAL
EDIT
TEST_OUTPUT: RANGE: 5V
EDIT
TEST_OUTPUT: AUTO CAL.:ON
EDIT
TEST_OUTPUT: AUTO CAL.:ON
ENTR EXIT
TEST_OUTPUT: AUTO CAL.:OFF
ENTR EXIT
EXIT
EXIT
EXIT
EXIT
EXIT
ENTR accepts
the new setting.
EXIT ignores the
new setting
06873B DCN6388