Teledyne API – Model T700 Dynamic Dilution Calibrator

Table 4-1: Calibrator Operating Modes

MODE
STANDBY
GENERATE
3
SETUP
1
This mode is not available in units without O
2
This mode is not available in units without internal photometers installed.
3
The revision of the Teledyne API software installed in this calibrator will be displayed following
the word SETUP. E.g. "SETUP G.4"
06873B DCN6388
The calibrator and all of its subsystems are inactive.
In this mode, the instrument is engaged in producing calibration gas
mixtures.
MANUAL In this mode, the instrument is engaged in producing calibration
gas mixtures.
PURGE
The calibrator is using diluent (zero air) to purge its internal
pneumatics of all source gas and previously created calibration
mixtures.
1
GPT
The calibrator is using the O
mix and generate calibration gas using the gas phase titration
method.
2
GPTPS
Stands for Gas Phase Titration Preset. In this mode the T700
determines the precise performance characteristics of the O
generator at the target values for an upcoming GPT calibration.
SETUP mode is being used to configure the calibrator.
DIAG
One of the calibrator's diagnostic modes is being utilized. When
the diagnostic functions that have the greatest potential to
conflict with generating concentrations are active, the instrument
is automatically placed into standby mode.
3
Overview of Operating Modes and Basic Operation
DESCRIPTION
generator and source gas inputs to
3
generators installed.
3
97