Figure 3-19: Set Up For T700 - Connecting The T700 To A Calibration Manifold - Teledyne T700 Operation Manual

Dynamic dilution calibrator
Hide thumbs Also See for T700:
Table of Contents

Advertisement

Teledyne API – Model T700 Dynamic Dilution Calibrator
Figure 3-19: Set up for T700 – Connecting the T700 to a Calibration Manifold
Note
06873B DCN6388
CONNECTING THE CALIBRATOR TO A CALIBRATION MANIFOLD
Using a calibration manifold provides a pneumatic interface between the calibration
system and other devices (or systems) which use the calibrator's gas output. Calibration
manifolds usually have one or more ports for connections to other external devices (such
as an analyzer).
• This method requires the analyzers connected to the calibration system have
internal zero/span valves.
• The manifold should be kept as clean as possible to avoid loss of sample gas
flow from blockages or constrictions.
• The manifolds as shown in the above drawing are oriented to simplify the
drawing.
Their actual orientation in your set-up is with the ports facing
upward. All unused ports must be capped.
• When initiating calibration, wait a minimum of 15 minutes for the calibrator to
flood the entire calibration manifold with calibration gas..
Getting Started
61

Advertisement

Table of Contents
loading

This manual is also suitable for:

T700u

Table of Contents