Internal Gas Pressure Sensors - Teledyne T700 Operation Manual

Dynamic dilution calibrator
Hide thumbs Also See for T700:
Table of Contents

Advertisement

Principles of Operation

10.2.2. INTERNAL GAS PRESSURE SENSORS

280
The T700 includes a single pressure regulator. Depending upon how many and which
options are installed in the T700 calibrator, there are between two and four pressure
sensors installed as well.
In the basic unit a printed circuit, assembly located near the front of the calibrator near
the MFCs includes sensors that measure the pressure of the diluent gas and the source
gas currently selected to flow into the calibrator. The calibrator monitors these sensors.
Should the pressure of one of them fall below 15 PSIG or rise above 36 PSIG a
warning is issued.
In units with the optional O
calibrator just behind the generator assembly includes a sensor that measures the gas
pressure of the zero air flowing into the generator. A regulator is also located on the gas
input to the O
generator that maintains the pressure differential needed for the critical
3
flow orifice to operate correctly.
Should the pressure of one of this sensor fall below 15 PSIG or rise above 25 PSIG
a warning is issued.
In calibrators with O
photometers installed, a second pressure located on the rear PCA
3
measures the pressure of gas in the photometer's absorption tube. This data is used by
the CPU when calculating the O
Teledyne API – Model T700 Dynamic Dilution Calibrator
generator installed a second PCA located at the rear of the
3
concentration inside the absorption tube.
3
06873B DCN6388

Advertisement

Table of Contents
loading

This manual is also suitable for:

T700u

Table of Contents