Table 9-11: Analog Output Test Function - Nominal Values Voltage Outputs - Teledyne T700 Operation Manual

Dynamic dilution calibrator
Hide thumbs Also See for T700:
Table of Contents

Advertisement

Teledyne API – Model T700 Dynamic Dilution Calibrator
Table 9-11: Analog Output Test Function – Nominal Values Voltage Outputs
STEP
1
2
3
4
5
6
06873B DCN6388
For each of the steps the output should be within 1% of the nominal value listed in the
table below except for the 0% step, which should be within 0mV ±2 to 3 mV. Ensure
you take into account any offset that may have been programmed into channel (See
Section 4.10.1.5).
FULL SCALE OUTPUT OF VOLTAGE RANGE
100MV
%
0
0
20
20 mV
40
40 mV
60
60 mV
80
80 mV
100
100 mV
If one or more of the steps fails to be within these ranges, it is likely that there has been
a failure of the either or both of the DACs and their associated circuitry on the
motherboard.
(see Section 4.10.1.3)
1V
NOMINAL OUTPUT VOLTAGE
0
0.2
0.4
0.6
0.8
1.0
Troubleshooting and Service
5V
10V
0
0
1
2
2
4
3
6
4
8
5
10
259