7.3.3.2. Setup Using a Calibration Manifold

Figure 7-4: External Photometer Validation Setup with Calibration Manifolds

Note

7.3.3.3. Calibration Manifold Exhaust/Vent Line

7.3.4. PERFORMING AN O
06873B DCN6388
Figure 7-4 shows the external zero air and O
photometer connected to the T700 Calibrator via calibration manifolds for both zero air
and O
.
3
The manifolds as shown in the above drawing are oriented to simplify the
drawing. The actual orientation in your setup is with the ports facing upward.
All unused ports should be capped. A Minimum of 1.1 LPM is required for the
external zero air source.
The manifold's excess gas should be vented to a suitable vent outside of the room. The
internal diameter of this vent should be large enough to avoid any appreciable pressure
drop, and it must be located sufficiently downstream of the output ports to ensure that no
ambient air enters the manifold due to eddy currents or back diffusion.
PHOTOMETER EXTERNAL CALIBRATION
3
The following procedure sets values held in the calibrator's memory for zero point
OFFSET and SLOPE.