O3 Photometer Gas Flow Calibration - Teledyne T700 Operation Manual

Dynamic dilution calibrator
Hide thumbs Also See for T700:
Table of Contents

Advertisement

Teledyne API – Model T700 Dynamic Dilution Calibrator

7.3.6. O3 PHOTOMETER GAS FLOW CALIBRATION

Note
06873B DCN6388
A separate flow meter is required for the procedure.
To calibrate the flow of gas through the T700 calibrator's optional photometer bench.
1. Turn OFF the T700 Dynamic Dilution Calibrator.
2. Attach the flow meter directly to the EXHAUST port of the T700 calibrator.
3. Turn the T700 Dynamic Dilution Calibrator ON.
4. Perform the following steps:
Calibration and Verification
209