Ozone Generator Temperature; Troubleshooting The Optional O Photometer; Dynamic Problems With The Optional O3 Photometer; Noisy Or Unstable O - Teledyne T700 Operation Manual

Dynamic dilution calibrator
Hide thumbs Also See for T700:
Table of Contents

Advertisement

Teledyne API – Model T700 Dynamic Dilution Calibrator

9.4.15.4. Ozone Generator Temperature

9.5. TROUBLESHOOTING THE OPTIONAL O

9.5.1. DYNAMIC PROBLEMS WITH THE OPTIONAL O3 PHOTOMETER

9.5.1.1. Noisy or Unstable O

06873B DCN6388
There are three possible causes for the Ozone Generator temperature to have failed.
The O
Gen heater has failed. Check the resistance between pins 5 and 6 on the
3
six-pin connector adjacent to the UV Lamp on the O
approximately 5 Ohms.
2
Assuming that the I
board, the FET Driver on the Relay Board may have failed. Using the
O3_GEN_HEATER parameter under the SIGNAL I/O submenu of the DIAG menu
as described above, turn the UV Lamp Heater on and off. Check the DC voltage
present between pin 1 and 2 on J14 of the Relay Board.
If the FET Driver has failed, there should be no change in the voltage across pins 1
and 2.
If the FET Driver checks out OK, the thermistor temperature sensor in the lamp
assembly may have failed. Unplug the connector to the Ozone Generator
Heater/Thermistor PCB, and measure the resistance of the thermistor between pins
5 and 6 of the 6-pin connector.
Dynamic problems are problems that only manifest themselves when the photometer is
measuring O
concentration gas mixtures. These can be the most difficult and time
3
consuming to isolate and resolve.
Since many photometer behaviors that appear to be a dynamic in nature are often a
symptom of a seemingly unrelated static problems, it is recommended that dynamic
problems not be addressed until all static problems, warning conditions and subsystems
have been checked and any problems found are resolved.
Once this has been accomplished, the following most common dynamic problems
should be checked.

Readings at Zero

3
Check for leaks in the pneumatic system as described in Section 8.2.1.
Confirm that the Zero gas is free of Ozone.
Confirm that the Source Lamp is fully inserted and that the lamp hold-down thumb-
screw is tight.
Check for a dirty Absorption Cell and/or pneumatic lines. Clean as necessary as
described in Section 8.2.2.
Disconnect the exhaust line from the optical bench (the pneumatic line at the lamp
end of the bench) and plug the port in the bench. If readings remain noisy, the
problem is in one of the electronic sections of the instrument. If readings become
quiet, the problem is in the instrument's pneumatics.
C bus is working and that there is no other failure with the Relay
Troubleshooting and Service
Generator. It should be
3
PHOTOMETER
3
265